Hello Time-Nuts,
I have an HP 5061A (relatively new chassis) with a very old Cs tube
installed (working) and I’m trying to wrap my head around the set up and
calibration of this unit, and just hoping for some help.
As a little background, HP changed the Zeeman modulation frequency, C-field
intensity, and A1 synthesizer frequency at serial number prefix 1336A. This
is detailed in CHANGE 20 (pages 7-9 and 7-10) Model 5061A Manual Changes in
the 5061A Manual, and covered in much more detail in HP Service Note
5061A-9 (document 05061-9061 6/80-02).
OLD NEW
Zeeman frequency 42.82kHz 53.53kHz
C-field 61 milligauss 76
milligauss
Synth. Frequency 12,631,771.6 Hz 12,631,772.5 Hz
I can put the tube on frequency at the new synthesizer frequency by
changing the C-field current (closer to the old C-field value, by changing
A15R19).
The Zeeman modulation frequency is a complete mystery to me. Is it
related to the value of the C-field that you are running?
If you are comparing the unit to GPS to calibrate it do you really care
what the Zeeman frequency is? Adjust the C-field to put it on frequency
and call it a day?
I’m not sure if there was a physical change in the Cs tube as well. But HP
mentions that the new values should have better stability.
Thanks for any insight.
Skip Withrow
Skip,
There is a great deal of expertise as you know here.
So several comments to get the ball rolling.
The zeeman frequency allows you to establish the optimal beam current main
peak by changing the magnetic field. What this is really doing is
optimizing the state of the cesium atoms. Its physically separating them
further apart so that when the beam hits the B magnet the bad cesiums head
off to the getter. Thus the best signal to noise possible. I believe it
also enhances focus of the beams. Less dispersion.
Now there is no doubt you can use GPS to set frequency and optimize the
beam current. But that particular point might not be the best signal to
noise possible. That is the way I have always set Frankenstein since beam
is impossible to see.
Your last question is the hardest. Just use GPS and be done.
I am really interested in the groups answer to that.
All corrections to my lack of knowledge are greatly appreciated.
Regards
Paul
WB8TSL
On Fri, Mar 17, 2017 at 11:37 AM, Skip Withrow skip.withrow@gmail.com
wrote:
Hello Time-Nuts,
I have an HP 5061A (relatively new chassis) with a very old Cs tube
installed (working) and I’m trying to wrap my head around the set up and
calibration of this unit, and just hoping for some help.
As a little background, HP changed the Zeeman modulation frequency, C-field
intensity, and A1 synthesizer frequency at serial number prefix 1336A.
This
is detailed in CHANGE 20 (pages 7-9 and 7-10) Model 5061A Manual Changes in
the 5061A Manual, and covered in much more detail in HP Service Note
5061A-9 (document 05061-9061 6/80-02).
OLD NEW
Zeeman frequency 42.82kHz 53.53kHz
C-field 61 milligauss 76
milligauss
Synth. Frequency 12,631,771.6 Hz 12,631,772.5 Hz
I can put the tube on frequency at the new synthesizer frequency by
changing the C-field current (closer to the old C-field value, by changing
A15R19).
The Zeeman modulation frequency is a complete mystery to me. Is it
related to the value of the C-field that you are running?
If you are comparing the unit to GPS to calibrate it do you really care
what the Zeeman frequency is? Adjust the C-field to put it on frequency
and call it a day?
I’m not sure if there was a physical change in the Cs tube as well. But HP
mentions that the new values should have better stability.
Thanks for any insight.
Skip Withrow
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.