My question is this. If I measured a 10 volt reference known to be good to
8 1/2 digits, but with the LSD being > 0, assuming a transfer standard with
traceability and documented uncertainties etc, would a 6 1/2 digit
voltmeter read 10.00000 volts, or round up to the limits of its AD
converter resolution; say X uV over 10.00000 volts?
I hope this is not too basic a question.
Russ
I very much enjoy what is shared here. It has led me down the path of
learning more about Metrology in general and the pointers to Flukecal site
from here are appreciated.
My question is this. If I measured a 10 volt reference known to be good to
8 1/2 digits, but with the LSD being > 0, assuming a transfer standard with
traceability and documented uncertainties etc, would a 6 1/2 digit
voltmeter read 10.00000 volts, or round up to the limits of its AD
converter resolution; say X uV over 10.00000 volts?
I hope this is not too basic a question.
Russ
MK
m k
Mon, Feb 20, 2017 6:57 PM
On 20/02/2017 18:46, Russ Ramirez wrote:
I very much enjoy what is shared here. It has led me down the path of
learning more about Metrology in general and the pointers to Flukecal site
from here are appreciated.
My question is this. If I measured a 10 volt reference known to be good to
8 1/2 digits, but with the LSD being > 0, assuming a transfer standard with
traceability and documented uncertainties etc, would a 6 1/2 digit
voltmeter read 10.00000 volts, or round up to the limits of its AD
converter resolution; say X uV over 10.00000 volts?
I hope this is not too basic a question.
Russ
Hi Russ,
some voltmeters will supply extra digits in the output they provide, but whether they are of any use depends upon how
stable the reference is in that particular 6.5 digit dvm, unfortunately it needs a dvm with better capability to prove
if those digits are believable.
On 20/02/2017 18:46, Russ Ramirez wrote:
> I very much enjoy what is shared here. It has led me down the path of
> learning more about Metrology in general and the pointers to Flukecal site
> from here are appreciated.
>
> My question is this. If I measured a 10 volt reference known to be good to
> 8 1/2 digits, but with the LSD being > 0, assuming a transfer standard with
> traceability and documented uncertainties etc, would a 6 1/2 digit
> voltmeter read 10.00000 volts, or round up to the limits of its AD
> converter resolution; say X uV over 10.00000 volts?
>
> I hope this is not too basic a question.
>
> Russ
> _______________________________________________
>
Hi Russ,
some voltmeters will supply extra digits in the output they provide, but whether they are of any use depends upon how
stable the reference is in that particular 6.5 digit dvm, unfortunately it needs a dvm with better capability to prove
if those digits are believable.
RR
Russ Ramirez
Mon, Feb 20, 2017 7:26 PM
My question is partly based on a real situation where I have 3 DMMs of
decent quality Keithley 2000, 34970A (34401A), and 3457A that all agree
with a good quality voltage reference that is exceeding its ppm/degree C
spec as far as I am able to tell. I felt that this was telling me that I
would need a 3458A to know how much closer to 10V the reference was at this
point. A quick calculation of 20 bits of ADC for example told me that I was
looking at 9.5 uV of resolution, or 0.6 uV at 24 bits noise free, etc. so
it did not seem to be entirely unreasonable to see the LSD of a 6 1/2 digit
DMM be higher than 1 or 2 counts.
Russ
On Mon, Feb 20, 2017 at 12:57 PM, m k m1k3k1@hotmail.com wrote:
On 20/02/2017 18:46, Russ Ramirez wrote:
I very much enjoy what is shared here. It has led me down the path of
learning more about Metrology in general and the pointers to Flukecal
from here are appreciated.
My question is this. If I measured a 10 volt reference known to be good
8 1/2 digits, but with the LSD being > 0, assuming a transfer standard
traceability and documented uncertainties etc, would a 6 1/2 digit
voltmeter read 10.00000 volts, or round up to the limits of its AD
converter resolution; say X uV over 10.00000 volts?
I hope this is not too basic a question.
Russ
Hi Russ,
some voltmeters will supply extra digits in the output they provide, but
whether they are of any use depends upon how
stable the reference is in that particular 6.5 digit dvm, unfortunately it
needs a dvm with better capability to prove
if those digits are believable.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
My question is partly based on a real situation where I have 3 DMMs of
decent quality Keithley 2000, 34970A (34401A), and 3457A that all agree
with a good quality voltage reference that is exceeding its ppm/degree C
spec as far as I am able to tell. I felt that this was telling me that I
would need a 3458A to know how much closer to 10V the reference was at this
point. A quick calculation of 20 bits of ADC for example told me that I was
looking at 9.5 uV of resolution, or 0.6 uV at 24 bits *noise free*, etc. so
it did not seem to be entirely unreasonable to see the LSD of a 6 1/2 digit
DMM be higher than 1 or 2 counts.
Russ
On Mon, Feb 20, 2017 at 12:57 PM, m k <m1k3k1@hotmail.com> wrote:
> On 20/02/2017 18:46, Russ Ramirez wrote:
> > I very much enjoy what is shared here. It has led me down the path of
> > learning more about Metrology in general and the pointers to Flukecal
> site
> > from here are appreciated.
> >
> > My question is this. If I measured a 10 volt reference known to be good
> to
> > 8 1/2 digits, but with the LSD being > 0, assuming a transfer standard
> with
> > traceability and documented uncertainties etc, would a 6 1/2 digit
> > voltmeter read 10.00000 volts, or round up to the limits of its AD
> > converter resolution; say X uV over 10.00000 volts?
> >
> > I hope this is not too basic a question.
> >
> > Russ
> > _______________________________________________
> >
> Hi Russ,
>
> some voltmeters will supply extra digits in the output they provide, but
> whether they are of any use depends upon how
> stable the reference is in that particular 6.5 digit dvm, unfortunately it
> needs a dvm with better capability to prove
> if those digits are believable.
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/volt-nuts
> and follow the instructions there.
>
PK
Poul-Henning Kamp
Mon, Feb 20, 2017 7:29 PM
My question is this. If I measured a 10 volt reference known to be good to
8 1/2 digits, but with the LSD being > 0, assuming a transfer standard with
traceability and documented uncertainties etc, would a 6 1/2 digit
voltmeter read 10.00000 volts, or round up to the limits of its AD
converter resolution; say X uV over 10.00000 volts?
Somewhere in your instrument there is an A/D converter of some kind.
Even if it is perfect, you can never know if the last bit from the A/D
represents a voltage which is slightly higher or slightly lower than
the number it reads.
This is why specifications for digital instruments always includes "± 1 digit"
In theory, if you have truly perfect electronics, it could be "± 0.5 digit",
because in that case you might as well use the same ±1 as everybody else
and double the size of the range.
The A/D in a 6½ digit instrument by definition must have at least
22 bits (2^21 = 2097152 + sign bit) and if we scale things so the
least significant bit is 5 microvolts, we have a full scale voltage
of 2^21 * 5e-6V = 10.48576V.
QED: Uncertainty on your 6½ digit meter on a 10V signal will always
be at least ± 5 microvolts.
If you do the same math for a true 8½ digit DVM, you need a 29 bit
A/D converter (2^28 = 268435456 + sign) which means you can use 40
nV as stepsize yielding 10.73V range, and thus in teory get ± 40nV
measurements.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
--------
In message <CAJgz9AYgQQjOmOK57yBiohF6Qh0DX6sEH-M_RxaBWQC54uDt7Q@mail.gmail.com>, Russ Ramirez writes:
>My question is this. If I measured a 10 volt reference known to be good to
>8 1/2 digits, but with the LSD being > 0, assuming a transfer standard with
>traceability and documented uncertainties etc, would a 6 1/2 digit
>voltmeter read 10.00000 volts, or round up to the limits of its AD
>converter resolution; say X uV over 10.00000 volts?
Somewhere in your instrument there is an A/D converter of some kind.
Even if it is perfect, you can never know if the last bit from the A/D
represents a voltage which is slightly higher or slightly lower than
the number it reads.
This is why specifications for digital instruments always includes "± 1 digit"
In theory, if you have *truly* perfect electronics, it could be "± 0.5 digit",
because in that case you might as well use the same ±1 as everybody else
and double the size of the range.
The A/D in a 6½ digit instrument by definition must have at least
22 bits (2^21 = 2097152 + sign bit) and if we scale things so the
least significant bit is 5 microvolts, we have a full scale voltage
of 2^21 * 5e-6V = 10.48576V.
QED: Uncertainty on your 6½ digit meter on a 10V signal will always
be at least ± 5 microvolts.
If you do the same math for a true 8½ digit DVM, you need a 29 bit
A/D converter (2^28 = 268435456 + sign) which means you can use 40
nV as stepsize yielding 10.73V range, and thus in teory get ± 40nV
measurements.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
TM
Tom Miller
Mon, Feb 20, 2017 7:33 PM
I believe you can get 7-1/2 digits out of the 3457A via the HPIB port.
Regards,
Tom
----- Original Message -----
From: "Russ Ramirez" russ.ramirez@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, February 20, 2017 2:26 PM
Subject: Re: [volt-nuts] Basic question concerning voltage references
My question is partly based on a real situation where I have 3 DMMs of
decent quality Keithley 2000, 34970A (34401A), and 3457A that all agree
with a good quality voltage reference that is exceeding its ppm/degree C
spec as far as I am able to tell. I felt that this was telling me that I
would need a 3458A to know how much closer to 10V the reference was at
this
point. A quick calculation of 20 bits of ADC for example told me that I
was
looking at 9.5 uV of resolution, or 0.6 uV at 24 bits noise free, etc.
so
it did not seem to be entirely unreasonable to see the LSD of a 6 1/2
digit
DMM be higher than 1 or 2 counts.
Russ
On Mon, Feb 20, 2017 at 12:57 PM, m k m1k3k1@hotmail.com wrote:
On 20/02/2017 18:46, Russ Ramirez wrote:
I very much enjoy what is shared here. It has led me down the path of
learning more about Metrology in general and the pointers to Flukecal
from here are appreciated.
My question is this. If I measured a 10 volt reference known to be good
8 1/2 digits, but with the LSD being > 0, assuming a transfer standard
traceability and documented uncertainties etc, would a 6 1/2 digit
voltmeter read 10.00000 volts, or round up to the limits of its AD
converter resolution; say X uV over 10.00000 volts?
I hope this is not too basic a question.
Russ
Hi Russ,
some voltmeters will supply extra digits in the output they provide, but
whether they are of any use depends upon how
stable the reference is in that particular 6.5 digit dvm, unfortunately
it
needs a dvm with better capability to prove
if those digits are believable.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
I believe you can get 7-1/2 digits out of the 3457A via the HPIB port.
Regards,
Tom
----- Original Message -----
From: "Russ Ramirez" <russ.ramirez@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Monday, February 20, 2017 2:26 PM
Subject: Re: [volt-nuts] Basic question concerning voltage references
> My question is partly based on a real situation where I have 3 DMMs of
> decent quality Keithley 2000, 34970A (34401A), and 3457A that all agree
> with a good quality voltage reference that is exceeding its ppm/degree C
> spec as far as I am able to tell. I felt that this was telling me that I
> would need a 3458A to know how much closer to 10V the reference was at
> this
> point. A quick calculation of 20 bits of ADC for example told me that I
> was
> looking at 9.5 uV of resolution, or 0.6 uV at 24 bits *noise free*, etc.
> so
> it did not seem to be entirely unreasonable to see the LSD of a 6 1/2
> digit
> DMM be higher than 1 or 2 counts.
>
> Russ
>
> On Mon, Feb 20, 2017 at 12:57 PM, m k <m1k3k1@hotmail.com> wrote:
>
>> On 20/02/2017 18:46, Russ Ramirez wrote:
>> > I very much enjoy what is shared here. It has led me down the path of
>> > learning more about Metrology in general and the pointers to Flukecal
>> site
>> > from here are appreciated.
>> >
>> > My question is this. If I measured a 10 volt reference known to be good
>> to
>> > 8 1/2 digits, but with the LSD being > 0, assuming a transfer standard
>> with
>> > traceability and documented uncertainties etc, would a 6 1/2 digit
>> > voltmeter read 10.00000 volts, or round up to the limits of its AD
>> > converter resolution; say X uV over 10.00000 volts?
>> >
>> > I hope this is not too basic a question.
>> >
>> > Russ
>> > _______________________________________________
>> >
>> Hi Russ,
>>
>> some voltmeters will supply extra digits in the output they provide, but
>> whether they are of any use depends upon how
>> stable the reference is in that particular 6.5 digit dvm, unfortunately
>> it
>> needs a dvm with better capability to prove
>> if those digits are believable.
>>
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>> mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
RR
Russ Ramirez
Mon, Feb 20, 2017 8:51 PM
Right, and also the 34970 using BenchView, though I could not find where
they claimed that. In the 34970/401 case I was simply seeing a lot of
variation in the LSD, and immediately discounted it as useful in any way. 6
1/2 digits had a close correlation (0.893 I think the number was) to
temperature, but 7 1/2 was all over the place and a relatively low value
for sigma.
Now you know how I spent part of last week and this weekend!
Russ
On Mon, Feb 20, 2017 at 1:33 PM, Tom Miller tmiller11147@verizon.net
wrote:
I believe you can get 7-1/2 digits out of the 3457A via the HPIB port.
Regards,
Tom
----- Original Message ----- From: "Russ Ramirez" russ.ramirez@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, February 20, 2017 2:26 PM
Subject: Re: [volt-nuts] Basic question concerning voltage references
My question is partly based on a real situation where I have 3 DMMs of
decent quality Keithley 2000, 34970A (34401A), and 3457A that all agree
with a good quality voltage reference that is exceeding its ppm/degree C
spec as far as I am able to tell. I felt that this was telling me that I
would need a 3458A to know how much closer to 10V the reference was at
this
point. A quick calculation of 20 bits of ADC for example told me that I
was
looking at 9.5 uV of resolution, or 0.6 uV at 24 bits noise free, etc.
so
it did not seem to be entirely unreasonable to see the LSD of a 6 1/2
digit
DMM be higher than 1 or 2 counts.
Russ
On Mon, Feb 20, 2017 at 12:57 PM, m k m1k3k1@hotmail.com wrote:
On 20/02/2017 18:46, Russ Ramirez wrote:
I very much enjoy what is shared here. It has led me down the path of
learning more about Metrology in general and the pointers to Flukecal
from here are appreciated.
My question is this. If I measured a 10 volt reference known to be good
8 1/2 digits, but with the LSD being > 0, assuming a transfer standard
traceability and documented uncertainties etc, would a 6 1/2 digit
voltmeter read 10.00000 volts, or round up to the limits of its AD
converter resolution; say X uV over 10.00000 volts?
I hope this is not too basic a question.
Russ
Hi Russ,
some voltmeters will supply extra digits in the output they provide, but
whether they are of any use depends upon how
stable the reference is in that particular 6.5 digit dvm, unfortunately
it
needs a dvm with better capability to prove
if those digits are believable.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
Right, and also the 34970 using BenchView, though I could not find where
they claimed that. In the 34970/401 case I was simply seeing a lot of
variation in the LSD, and immediately discounted it as useful in any way. 6
1/2 digits had a close correlation (0.893 I think the number was) to
temperature, but 7 1/2 was all over the place and a relatively low value
for sigma.
Now you know how I spent part of last week and this weekend!
Russ
On Mon, Feb 20, 2017 at 1:33 PM, Tom Miller <tmiller11147@verizon.net>
wrote:
> I believe you can get 7-1/2 digits out of the 3457A via the HPIB port.
>
> Regards,
> Tom
>
> ----- Original Message ----- From: "Russ Ramirez" <russ.ramirez@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Monday, February 20, 2017 2:26 PM
> Subject: Re: [volt-nuts] Basic question concerning voltage references
>
>
>
> My question is partly based on a real situation where I have 3 DMMs of
>> decent quality Keithley 2000, 34970A (34401A), and 3457A that all agree
>> with a good quality voltage reference that is exceeding its ppm/degree C
>> spec as far as I am able to tell. I felt that this was telling me that I
>> would need a 3458A to know how much closer to 10V the reference was at
>> this
>> point. A quick calculation of 20 bits of ADC for example told me that I
>> was
>> looking at 9.5 uV of resolution, or 0.6 uV at 24 bits *noise free*, etc.
>> so
>> it did not seem to be entirely unreasonable to see the LSD of a 6 1/2
>> digit
>> DMM be higher than 1 or 2 counts.
>>
>> Russ
>>
>> On Mon, Feb 20, 2017 at 12:57 PM, m k <m1k3k1@hotmail.com> wrote:
>>
>> On 20/02/2017 18:46, Russ Ramirez wrote:
>>> > I very much enjoy what is shared here. It has led me down the path of
>>> > learning more about Metrology in general and the pointers to Flukecal
>>> site
>>> > from here are appreciated.
>>> >
>>> > My question is this. If I measured a 10 volt reference known to be good
>>> to
>>> > 8 1/2 digits, but with the LSD being > 0, assuming a transfer standard
>>> with
>>> > traceability and documented uncertainties etc, would a 6 1/2 digit
>>> > voltmeter read 10.00000 volts, or round up to the limits of its AD
>>> > converter resolution; say X uV over 10.00000 volts?
>>> >
>>> > I hope this is not too basic a question.
>>> >
>>> > Russ
>>> > _______________________________________________
>>> >
>>> Hi Russ,
>>>
>>> some voltmeters will supply extra digits in the output they provide, but
>>> whether they are of any use depends upon how
>>> stable the reference is in that particular 6.5 digit dvm, unfortunately
>>> it
>>> needs a dvm with better capability to prove
>>> if those digits are believable.
>>>
>>>
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>>> mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>>>
>>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/m
>> ailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/m
> ailman/listinfo/volt-nuts
> and follow the instructions there.
>
JH
Jerry Hancock
Mon, Feb 20, 2017 8:53 PM
You can get 7.5 digits using the HIRES register over GPIB or by using the stat function from the front panel.
On Feb 20, 2017, at 11:33 AM, Tom Miller tmiller11147@verizon.net wrote:
I believe you can get 7-1/2 digits out of the 3457A via the HPIB port.
Regards,
Tom
----- Original Message ----- From: "Russ Ramirez" russ.ramirez@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, February 20, 2017 2:26 PM
Subject: Re: [volt-nuts] Basic question concerning voltage references
My question is partly based on a real situation where I have 3 DMMs of
decent quality Keithley 2000, 34970A (34401A), and 3457A that all agree
with a good quality voltage reference that is exceeding its ppm/degree C
spec as far as I am able to tell. I felt that this was telling me that I
would need a 3458A to know how much closer to 10V the reference was at this
point. A quick calculation of 20 bits of ADC for example told me that I was
looking at 9.5 uV of resolution, or 0.6 uV at 24 bits noise free, etc. so
it did not seem to be entirely unreasonable to see the LSD of a 6 1/2 digit
DMM be higher than 1 or 2 counts.
Russ
On Mon, Feb 20, 2017 at 12:57 PM, m k m1k3k1@hotmail.com wrote:
On 20/02/2017 18:46, Russ Ramirez wrote:
I very much enjoy what is shared here. It has led me down the path of
learning more about Metrology in general and the pointers to Flukecal
from here are appreciated.
My question is this. If I measured a 10 volt reference known to be good
8 1/2 digits, but with the LSD being > 0, assuming a transfer standard
traceability and documented uncertainties etc, would a 6 1/2 digit
voltmeter read 10.00000 volts, or round up to the limits of its AD
converter resolution; say X uV over 10.00000 volts?
I hope this is not too basic a question.
Russ
Hi Russ,
some voltmeters will supply extra digits in the output they provide, but
whether they are of any use depends upon how
stable the reference is in that particular 6.5 digit dvm, unfortunately it
needs a dvm with better capability to prove
if those digits are believable.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
You can get 7.5 digits using the HIRES register over GPIB or by using the stat function from the front panel.
> On Feb 20, 2017, at 11:33 AM, Tom Miller <tmiller11147@verizon.net> wrote:
>
> I believe you can get 7-1/2 digits out of the 3457A via the HPIB port.
>
> Regards,
> Tom
>
> ----- Original Message ----- From: "Russ Ramirez" <russ.ramirez@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Monday, February 20, 2017 2:26 PM
> Subject: Re: [volt-nuts] Basic question concerning voltage references
>
>
>> My question is partly based on a real situation where I have 3 DMMs of
>> decent quality Keithley 2000, 34970A (34401A), and 3457A that all agree
>> with a good quality voltage reference that is exceeding its ppm/degree C
>> spec as far as I am able to tell. I felt that this was telling me that I
>> would need a 3458A to know how much closer to 10V the reference was at this
>> point. A quick calculation of 20 bits of ADC for example told me that I was
>> looking at 9.5 uV of resolution, or 0.6 uV at 24 bits *noise free*, etc. so
>> it did not seem to be entirely unreasonable to see the LSD of a 6 1/2 digit
>> DMM be higher than 1 or 2 counts.
>>
>> Russ
>>
>> On Mon, Feb 20, 2017 at 12:57 PM, m k <m1k3k1@hotmail.com> wrote:
>>
>>> On 20/02/2017 18:46, Russ Ramirez wrote:
>>> > I very much enjoy what is shared here. It has led me down the path of
>>> > learning more about Metrology in general and the pointers to Flukecal
>>> site
>>> > from here are appreciated.
>>> >
>>> > My question is this. If I measured a 10 volt reference known to be good
>>> to
>>> > 8 1/2 digits, but with the LSD being > 0, assuming a transfer standard
>>> with
>>> > traceability and documented uncertainties etc, would a 6 1/2 digit
>>> > voltmeter read 10.00000 volts, or round up to the limits of its AD
>>> > converter resolution; say X uV over 10.00000 volts?
>>> >
>>> > I hope this is not too basic a question.
>>> >
>>> > Russ
>>> > _______________________________________________
>>> >
>>> Hi Russ,
>>>
>>> some voltmeters will supply extra digits in the output they provide, but
>>> whether they are of any use depends upon how
>>> stable the reference is in that particular 6.5 digit dvm, unfortunately it
>>> needs a dvm with better capability to prove
>>> if those digits are believable.
>>>
>>>
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>>> mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
RR
Russ Ramirez
Mon, Feb 20, 2017 9:13 PM
Thanks Jerry, I'll give that a try later just to see what it does :-)
Russ
On Mon, Feb 20, 2017 at 2:53 PM, Jerry Hancock jerry@hanler.com wrote:
You can get 7.5 digits using the HIRES register over GPIB or by using the
stat function from the front panel.
I believe you can get 7-1/2 digits out of the 3457A via the HPIB port.
Regards,
Tom
----- Original Message ----- From: "Russ Ramirez" <
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, February 20, 2017 2:26 PM
Subject: Re: [volt-nuts] Basic question concerning voltage references
My question is partly based on a real situation where I have 3 DMMs of
decent quality Keithley 2000, 34970A (34401A), and 3457A that all agree
with a good quality voltage reference that is exceeding its ppm/degree C
spec as far as I am able to tell. I felt that this was telling me that I
would need a 3458A to know how much closer to 10V the reference was at
point. A quick calculation of 20 bits of ADC for example told me that I
looking at 9.5 uV of resolution, or 0.6 uV at 24 bits noise free,
it did not seem to be entirely unreasonable to see the LSD of a 6 1/2
DMM be higher than 1 or 2 counts.
Russ
On Mon, Feb 20, 2017 at 12:57 PM, m k m1k3k1@hotmail.com wrote:
On 20/02/2017 18:46, Russ Ramirez wrote:
I very much enjoy what is shared here. It has led me down the path of
learning more about Metrology in general and the pointers to Flukecal
from here are appreciated.
My question is this. If I measured a 10 volt reference known to be
8 1/2 digits, but with the LSD being > 0, assuming a transfer
traceability and documented uncertainties etc, would a 6 1/2 digit
voltmeter read 10.00000 volts, or round up to the limits of its AD
converter resolution; say X uV over 10.00000 volts?
I hope this is not too basic a question.
Russ
Hi Russ,
some voltmeters will supply extra digits in the output they provide,
whether they are of any use depends upon how
stable the reference is in that particular 6.5 digit dvm,
mailman/listinfo/volt-nuts
and follow the instructions there.
mailman/listinfo/volt-nuts
and follow the instructions there.
Thanks Jerry, I'll give that a try later just to see what it does :-)
Russ
On Mon, Feb 20, 2017 at 2:53 PM, Jerry Hancock <jerry@hanler.com> wrote:
> You can get 7.5 digits using the HIRES register over GPIB or by using the
> stat function from the front panel.
>
>
> > On Feb 20, 2017, at 11:33 AM, Tom Miller <tmiller11147@verizon.net>
> wrote:
> >
> > I believe you can get 7-1/2 digits out of the 3457A via the HPIB port.
> >
> > Regards,
> > Tom
> >
> > ----- Original Message ----- From: "Russ Ramirez" <
> russ.ramirez@gmail.com>
> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > Sent: Monday, February 20, 2017 2:26 PM
> > Subject: Re: [volt-nuts] Basic question concerning voltage references
> >
> >
> >> My question is partly based on a real situation where I have 3 DMMs of
> >> decent quality Keithley 2000, 34970A (34401A), and 3457A that all agree
> >> with a good quality voltage reference that is exceeding its ppm/degree C
> >> spec as far as I am able to tell. I felt that this was telling me that I
> >> would need a 3458A to know how much closer to 10V the reference was at
> this
> >> point. A quick calculation of 20 bits of ADC for example told me that I
> was
> >> looking at 9.5 uV of resolution, or 0.6 uV at 24 bits *noise free*,
> etc. so
> >> it did not seem to be entirely unreasonable to see the LSD of a 6 1/2
> digit
> >> DMM be higher than 1 or 2 counts.
> >>
> >> Russ
> >>
> >> On Mon, Feb 20, 2017 at 12:57 PM, m k <m1k3k1@hotmail.com> wrote:
> >>
> >>> On 20/02/2017 18:46, Russ Ramirez wrote:
> >>> > I very much enjoy what is shared here. It has led me down the path of
> >>> > learning more about Metrology in general and the pointers to Flukecal
> >>> site
> >>> > from here are appreciated.
> >>> >
> >>> > My question is this. If I measured a 10 volt reference known to be
> good
> >>> to
> >>> > 8 1/2 digits, but with the LSD being > 0, assuming a transfer
> standard
> >>> with
> >>> > traceability and documented uncertainties etc, would a 6 1/2 digit
> >>> > voltmeter read 10.00000 volts, or round up to the limits of its AD
> >>> > converter resolution; say X uV over 10.00000 volts?
> >>> >
> >>> > I hope this is not too basic a question.
> >>> >
> >>> > Russ
> >>> > _______________________________________________
> >>> >
> >>> Hi Russ,
> >>>
> >>> some voltmeters will supply extra digits in the output they provide,
> but
> >>> whether they are of any use depends upon how
> >>> stable the reference is in that particular 6.5 digit dvm,
> unfortunately it
> >>> needs a dvm with better capability to prove
> >>> if those digits are believable.
> >>>
> >>>
> >>> _______________________________________________
> >>> volt-nuts mailing list -- volt-nuts@febo.com
> >>> To unsubscribe, go to https://www.febo.com/cgi-bin/
> >>> mailman/listinfo/volt-nuts
> >>> and follow the instructions there.
> >>>
> >> _______________________________________________
> >> volt-nuts mailing list -- volt-nuts@febo.com
> >> To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/volt-nuts
> >> and follow the instructions there.
> >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/volt-nuts
> and follow the instructions there.
>
D
David
Tue, Feb 21, 2017 2:39 AM
On Mon, 20 Feb 2017 19:29:55 +0000, you wrote:
...
The A/D in a 6½ digit instrument by definition must have at least
22 bits (2^21 = 2097152 + sign bit) and if we scale things so the
least significant bit is 5 microvolts, we have a full scale voltage
of 2^21 * 5e-6V = 10.48576V.
QED: Uncertainty on your 6½ digit meter on a 10V signal will always
be at least ± 5 microvolts.
If you do the same math for a true 8½ digit DVM, you need a 29 bit
A/D converter (2^28 = 268435456 + sign) which means you can use 40
nV as stepsize yielding 10.73V range, and thus in teory get ± 40nV
measurements.
I just have a minor quibble with the above which does not alter your
point.
There is no requirement for an ADC output to be a power of 2. Slope
integration and voltage-to-frequency designs commonly use or used BCD
counters to produce a power of 10 output.
I remember when supressing negative zero was a new feature.
On Mon, 20 Feb 2017 19:29:55 +0000, you wrote:
>...
>
>The A/D in a 6½ digit instrument by definition must have at least
>22 bits (2^21 = 2097152 + sign bit) and if we scale things so the
>least significant bit is 5 microvolts, we have a full scale voltage
>of 2^21 * 5e-6V = 10.48576V.
>
>QED: Uncertainty on your 6½ digit meter on a 10V signal will always
>be at least ± 5 microvolts.
>
>If you do the same math for a true 8½ digit DVM, you need a 29 bit
>A/D converter (2^28 = 268435456 + sign) which means you can use 40
>nV as stepsize yielding 10.73V range, and thus in teory get ± 40nV
>measurements.
I just have a minor quibble with the above which does not alter your
point.
There is no requirement for an ADC output to be a power of 2. Slope
integration and voltage-to-frequency designs commonly use or used BCD
counters to produce a power of 10 output.
I remember when supressing negative zero was a new feature.
A
acbern@gmx.de
Tue, Feb 21, 2017 10:34 AM
If you want to do a comparison measurement with consideration of the actual measurement uncertainty, then what is relevant is not the theoretical resolution of the meter but what the meter data sheet tells you with respect of that matter. It is ususlly called the transfer accuracy. Applies when you do two measurements with one meter, at similar voltages. E.g. the drift of a reference after it has been exposed to a temperature change. It takes into account also things like differential non-linearity of the A/D, short term noise/flicker/drift issues and so on.
For a 3458A at 10V, it is 0.05+0.05ppm (range and reading), so 0,1ppm (0.2ppm for two measurements in total). Now for 6.5 digit DMMs it is obviously less, and not always specified. As an alternative, if the value is not given, one can take the uncertainty referenced to the range (see Fluke 'Calibration: Philosphy in Practice'). The 34401 has the A/D linearity specified, it is 3ppm. That helps as well. No way arround analyzing such transfer measurement uncertainty for any given meter based on the data sheet values.
cheers
If you want to do a comparison measurement with consideration of the actual measurement uncertainty, then what is relevant is not the theoretical resolution of the meter but what the meter data sheet tells you with respect of that matter. It is ususlly called the transfer accuracy. Applies when you do two measurements with one meter, at similar voltages. E.g. the drift of a reference after it has been exposed to a temperature change. It takes into account also things like differential non-linearity of the A/D, short term noise/flicker/drift issues and so on.
For a 3458A at 10V, it is 0.05+0.05ppm (range and reading), so 0,1ppm (0.2ppm for two measurements in total). Now for 6.5 digit DMMs it is obviously less, and not always specified. As an alternative, if the value is not given, one can take the uncertainty referenced to the range (see Fluke 'Calibration: Philosphy in Practice'). The 34401 has the A/D linearity specified, it is 3ppm. That helps as well. No way arround analyzing such transfer measurement uncertainty for any given meter based on the data sheet values.
cheers