Hello.
I repeated the experiment with a better setup, I also added some alloys
that are already arrived. So far, a really "low" EMF solder hasn't been
identified.
Measurements have been done between water ice point and boiling point with
Agilent 34401A.
I don't expect the curves be linear, but here it's assumed they are.
Copper - Sn96.5/Ag3/Cu0.5 -> 3.35uV/K
Copper - Sn95.5/Ag3.8/Cu0.7 -> 3.22uV/K
Copper - Sn60/Pb40 -> 3.34uV/K
Copper - Pb92.5/Sn5/Ag2.5 -> 3.02uV/K
Copper - Brass -> 3.30uV/K
I am waiting for Sn96/Ag4, Sn99/Cu1.
Best regards,
Andrea Baldoni
Hello,
I do not know exactly how the Seebeck coefficients are "mixed" within a
alloy.
Silver and copper have near equal coefficients.
Sn should be similar to Pb on the same side of Cu.
On the other side of Cu there are Cd (no longer allowed for ROHS) and Sb
(Antimony) and Ge (Germanium).
So I would go for Sb or Ge doped Tin solders if you do not want to have
100% Ag + Cu
With best regards
Andreas
Am 03.07.2016 um 12:33 schrieb Andrea Baldoni:
Hello.
I repeated the experiment with a better setup, I also added some alloys
that are already arrived. So far, a really "low" EMF solder hasn't been
identified.
Measurements have been done between water ice point and boiling point with
Agilent 34401A.
I don't expect the curves be linear, but here it's assumed they are.
Copper - Sn96.5/Ag3/Cu0.5 -> 3.35uV/K
Copper - Sn95.5/Ag3.8/Cu0.7 -> 3.22uV/K
Copper - Sn60/Pb40 -> 3.34uV/K
Copper - Pb92.5/Sn5/Ag2.5 -> 3.02uV/K
Copper - Brass -> 3.30uV/K
I am waiting for Sn96/Ag4, Sn99/Cu1.
Best regards,
Andrea Baldoni
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Wikipedia cites SN10 PB88 AG02 as having low thermal EMF properties. There are several listings for it on Ebay at the moment. Although not the same as Sn18 solder, it's pretty close. I'd think that the thermal properties would be pretty close as well.
Dave M
----- Original Message -----
From: "Andreas Jahn" Andreas_-_Jahn@t-online.de
To: volt-nuts@febo.com
Sent: Sunday, July 3, 2016 10:14:16 AM
Subject: Re: [volt-nuts] Thermal EMF - more results
Hello,
I do not know exactly how the Seebeck coefficients are "mixed" within a
alloy.
Silver and copper have near equal coefficients.
Sn should be similar to Pb on the same side of Cu.
On the other side of Cu there are Cd (no longer allowed for ROHS) and Sb
(Antimony) and Ge (Germanium).
So I would go for Sb or Ge doped Tin solders if you do not want to have
100% Ag + Cu
With best regards
Andreas
Am 03.07.2016 um 12:33 schrieb Andrea Baldoni:
Hello.
I repeated the experiment with a better setup, I also added some alloys
that are already arrived. So far, a really "low" EMF solder hasn't been
identified.
Measurements have been done between water ice point and boiling point with
Agilent 34401A.
I don't expect the curves be linear, but here it's assumed they are.
Copper - Sn96.5/Ag3/Cu0.5 -> 3.35uV/K
Copper - Sn95.5/Ag3.8/Cu0.7 -> 3.22uV/K
Copper - Sn60/Pb40 -> 3.34uV/K
Copper - Pb92.5/Sn5/Ag2.5 -> 3.02uV/K
Copper - Brass -> 3.30uV/K
I am waiting for Sn96/Ag4, Sn99/Cu1.
Best regards,
Andrea Baldoni
istinfo/volt-nuts
and follow the instructions there.
On Sun, Jul 03, 2016 at 12:33:04PM +0200, Andrea Baldoni wrote:
Hello.
I repeated the experiment with a better setup, I also added
some alloys that are already arrived. So far, a really "low"
EMF solder hasn't been identified.
Measurements have been done between water ice point and boiling
point with Agilent 34401A. I don't expect the curves be linear,
but here it's assumed they are.
Copper - Sn96.5/Ag3/Cu0.5 -> 3.35uV/K
Copper - Sn95.5/Ag3.8/Cu0.7 -> 3.22uV/K
Copper - Sn60/Pb40 -> 3.34uV/K
Copper - Pb92.5/Sn5/Ag2.5 -> 3.02uV/K
Copper - Brass -> 3.30uV/K
I am waiting for Sn96/Ag4, Sn99/Cu1.
Any plans on testing Sn42/Bi57.6/Ag0.4 ?
Thanks,
Herbert
Best regards,
Andrea Baldoni
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
On Wed, Jul 06, 2016 at 03:42:18PM +0200, Herbert Poetzl wrote:
Any plans on testing Sn42/Bi57.6/Ag0.4 ?
Hello Herbert.
I don't have any source of it. My supplier (Heraeus) hasn't it, at least,
it wasn't included in the list of samples they are able to give me on request.
If you have it and you could ship it (a meter of 0.5-1mm wire or so would be
sufficient for me and I will give it back after), I will be happy to test it.
Today I received the Sn99.3/Cu0.7 and I will post result soon.
Best regards,
Andrea Baldoni
so far so good.
but what do we learn from this? the solder alloy is essentially irrelevant
Gesendet: Mittwoch, 06. Juli 2016 um 15:42 Uhr
Von: "Herbert Poetzl" herbert@13thfloor.at
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] Thermal EMF - more results
On Sun, Jul 03, 2016 at 12:33:04PM +0200, Andrea Baldoni wrote:
Hello.
I repeated the experiment with a better setup, I also added
some alloys that are already arrived. So far, a really "low"
EMF solder hasn't been identified.
Measurements have been done between water ice point and boiling
point with Agilent 34401A. I don't expect the curves be linear,
but here it's assumed they are.
Copper - Sn96.5/Ag3/Cu0.5 -> 3.35uV/K
Copper - Sn95.5/Ag3.8/Cu0.7 -> 3.22uV/K
Copper - Sn60/Pb40 -> 3.34uV/K
Copper - Pb92.5/Sn5/Ag2.5 -> 3.02uV/K
Copper - Brass -> 3.30uV/K
I am waiting for Sn96/Ag4, Sn99/Cu1.
Any plans on testing Sn42/Bi57.6/Ag0.4 ?
Thanks,
Herbert
Best regards,
Andrea Baldoni
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
On Wed, Jul 06, 2016 at 08:53:19PM +0200, Andrea Baldoni wrote:
On Wed, Jul 06, 2016 at 03:42:18PM +0200, Herbert Poetzl wrote:
Any plans on testing Sn42/Bi57.6/Ag0.4 ?
Hello Herbert.
Hey Andrea,
I don't have any source of it. My supplier (Heraeus) hasn't it,
at least, it wasn't included in the list of samples they are
able to give me on request.
If you have it and you could ship it (a meter of 0.5-1mm wire
or so would be sufficient for me and I will give it back
after), I will be happy to test it.
Send me your address (off list) and I'll send you a meter for
testing.
Best,
Herbert
Today I received the Sn99.3/Cu0.7 and I will post result soon.
Best regards,
Andrea Baldoni
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
All this talk about thermal EMFs may not be particularly relevant.
Solder is used in connections between two metallic objects, like wire and
a terminal. The thermal EMFs between a soldered copper wire and copper terminal
have exactly equal and opposite values between the solder and the metals.
I first met this in mid 60s making thermocouples, I was told that they should be welded
not soldered because of the solder metal EMFs. I did both soldered and welded
over the years, mainly for practical considerations, and although some situations
were quite critical, I have never found evidence that solder matters. Even if you consider
a metal like tin diffusing into the copper and changing its EMF, adjacent to the area
of diffusion is a reverse gradient of tin concentration and thermodynamic considerations
dictate that there is no effect for static conditions.
The worst effect is the response of the junction to a persistent thermal
gradient, but this is hard to maintain over the distance of solder thickness.
cheers,
Neville Michie
On 9 Jul 2016, at 8:56 am, Herbert Poetzl herbert@13thfloor.at wrote:
On Wed, Jul 06, 2016 at 08:53:19PM +0200, Andrea Baldoni wrote:
On Wed, Jul 06, 2016 at 03:42:18PM +0200, Herbert Poetzl wrote:
Any plans on testing Sn42/Bi57.6/Ag0.4 ?
Hello Herbert.
Hey Andrea,
I don't have any source of it. My supplier (Heraeus) hasn't it,
at least, it wasn't included in the list of samples they are
able to give me on request.
If you have it and you could ship it (a meter of 0.5-1mm wire
or so would be sufficient for me and I will give it back
after), I will be happy to test it.
Send me your address (off list) and I'll send you a meter for
testing.
Best,
Herbert
Today I received the Sn99.3/Cu0.7 and I will post result soon.
Best regards,
Andrea Baldoni
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
I will argue that this issue is extremely relevant, especially in low voltage and low resistance measurements. You are correct in your position if the entire junction is in thermal stability. If however, the junction is exposed to a sudden change in temperature, air drafts being the most likely, then your example of copper wire soldered to copper terminal will indeed exhibit thermal EMF due to the difference in thermal mass of the copper wire to solder vs. the solder to copper thermal. Air drafts in the metrology lab is typically the concern in thermal EMF. A simple oscillating fan on the interconnect terminals during a precise, low (1 to 10) ohms, 4-wire, 1 to .01 PPM measurement will convince you. It is also easier to see if graphing the measurement results. The generated thermal EMF increases as the difference in thermal mass increases between the soldered principles in the junction. A #12 copper wire soldered to another #12 copper wire - no concern, even in a draft. A #12 copper wire soldered to a #22 copper wire - no concern if at thermal stability but blow on it and observe.
Regards,
Don
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Neville Michie
Sent: Friday, July 08, 2016 7:08 PM
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: Re: [volt-nuts] Thermal EMF - more results
All this talk about thermal EMFs may not be particularly relevant.
Solder is used in connections between two metallic objects, like wire and a terminal. The thermal EMFs between a soldered copper wire and copper terminal have exactly equal and opposite values between the solder and the metals.
I first met this in mid 60s making thermocouples, I was told that they should be welded not soldered because of the solder metal EMFs. I did both soldered and welded over the years, mainly for practical considerations, and although some situations were quite critical, I have never found evidence that solder matters. Even if you consider a metal like tin diffusing into the copper and changing its EMF, adjacent to the area of diffusion is a reverse gradient of tin concentration and thermodynamic considerations dictate that there is no effect for static conditions.
The worst effect is the response of the junction to a persistent thermal gradient, but this is hard to maintain over the distance of solder thickness.
cheers,
Neville Michie
On 9 Jul 2016, at 8:56 am, Herbert Poetzl herbert@13thfloor.at wrote:
On Wed, Jul 06, 2016 at 08:53:19PM +0200, Andrea Baldoni wrote:
On Wed, Jul 06, 2016 at 03:42:18PM +0200, Herbert Poetzl wrote:
Any plans on testing Sn42/Bi57.6/Ag0.4 ?
Hello Herbert.
Hey Andrea,
I don't have any source of it. My supplier (Heraeus) hasn't it, at
least, it wasn't included in the list of samples they are able to
give me on request.
If you have it and you could ship it (a meter of 0.5-1mm wire or so
would be sufficient for me and I will give it back after), I will be
happy to test it.
Send me your address (off list) and I'll send you a meter for testing.
Best,
Herbert
Today I received the Sn99.3/Cu0.7 and I will post result soon.
Best regards,
Andrea Baldoni
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.