I have been testing two 3458As against a known good Fluke 732A. Each 3458A
was calibrated for DCV before the start of the measurements so they both
started at the same point. I have been running a series of tests
consisting of measuring the Cal Constant as detailed in Service Note 18 for
each meter several times a day (to calculate the drift per the procedure in
SN 18). I also measure the 732A voltage in each meter each time using NLPC
100 and NRDGS 100 and then recording the STDEV, MEAN, MAX, and MIN values.
What I have observed is that the Cal Constant is acceptably low but the
MEAN value per measurement is drifting up in one meter and drifting down
in the other. The unit drifting down has a new A3 board installed and the
unit drifting upward is an Agilent 3458A only a few years old so would not
be expected to have a drifting A3 board. It was only calibrated 1 time per
the internal REV number, so would have likely been re-calibrated if the A3
board was replaced.
My question is what is the likely cause of the drift in the MEAN voltage
reading if the Cal Constant value is relatively constant? SN 18 says the
drift rate of the Cal Constant is an indication of drift in the A3 AD
board, but I believe it assumes the A9 voltage ref board has a constant
value over time. If this is not true (e;g., the voltage ref is drifting),
I think this would explain the drift in the voltage reading even though the
Cal Constant is relatively constant. Any opinions on this?
Thanks,
Randy Evans
What you describe (A9 drift) would be the explanation. However question is how much you see. The A9 should not drift much in the intervals you talk about (days).
Also, you should run the ACAL to determine the cal constant not several times a day but with several days inbetween, and then divide by number of days to determin drift
Frequent ACAL may not give you good results (random fluctuations such as noise, temp... and their impact during cal gets higher the shorter the time between ACALs)
Gesendet: Donnerstag, 30. November 2017 um 07:40 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] Drifting 3458As
I have been testing two 3458As against a known good Fluke 732A. Each 3458A
was calibrated for DCV before the start of the measurements so they both
started at the same point. I have been running a series of tests
consisting of measuring the Cal Constant as detailed in Service Note 18 for
each meter several times a day (to calculate the drift per the procedure in
SN 18). I also measure the 732A voltage in each meter each time using NLPC
100 and NRDGS 100 and then recording the STDEV, MEAN, MAX, and MIN values.
What I have observed is that the Cal Constant is acceptably low but the
MEAN value per measurement is drifting up in one meter and drifting down
in the other. The unit drifting down has a new A3 board installed and the
unit drifting upward is an Agilent 3458A only a few years old so would not
be expected to have a drifting A3 board. It was only calibrated 1 time per
the internal REV number, so would have likely been re-calibrated if the A3
board was replaced.
My question is what is the likely cause of the drift in the MEAN voltage
reading if the Cal Constant value is relatively constant? SN 18 says the
drift rate of the Cal Constant is an indication of drift in the A3 AD
board, but I believe it assumes the A9 voltage ref board has a constant
value over time. If this is not true (e;g., the voltage ref is drifting),
I think this would explain the drift in the voltage reading even though the
Cal Constant is relatively constant. Any opinions on this?
Thanks,
Randy Evans
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
I am running the ACAL to check cal constant drift over several times a day
and over a 7 day period (per SN18). The varying times of the day are to
see how temperature sensitivity affects the cal constant drift since the
room temperature is not constant through the day but is relatively constant
at the same time of day each day; i.e., the furnace is off during the night
so mornings are about 5 degrees C cooler than the afternoon. Even after
ACAL, I am seeing up to an absolute difference of 3.5 uV on the 732A
voltage reading between the HP and Agilent 3458As, depending on the time of
day. I suspect the A9 board in the Agilent is the reason since the HP
3458A reads much closer to the 732A value than the Agilent 3458A over
temperature. Based on long term comparisons with other 732As and a 732B, I
am confident the 732A is stable within 0.1uV.
I am using an ebay A9 board in the Agilent 3458A for the current testing
since I suspect the original A9 board in the Agilent 3458A. I am
separately testing the original A9 board since I am not completely
confident in the ebay A9 board. I do not have enough data to draw a
conclusion yet about the original A9 board. My current concern is to see
if there is agreement that the A9 board is likely the culprit.
Thanks,
Randy Evans
On Fri, Dec 1, 2017 at 1:58 AM, acbern@gmx.de wrote:
What you describe (A9 drift) would be the explanation. However question is
how much you see. The A9 should not drift much in the intervals you talk
about (days).
Also, you should run the ACAL to determine the cal constant not several
times a day but with several days inbetween, and then divide by number of
days to determin drift
Frequent ACAL may not give you good results (random fluctuations such as
noise, temp... and their impact during cal gets higher the shorter the time
between ACALs)
Gesendet: Donnerstag, 30. November 2017 um 07:40 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] Drifting 3458As
I have been testing two 3458As against a known good Fluke 732A. Each
3458A
was calibrated for DCV before the start of the measurements so they both
started at the same point. I have been running a series of tests
consisting of measuring the Cal Constant as detailed in Service Note 18
for
each meter several times a day (to calculate the drift per the procedure
in
SN 18). I also measure the 732A voltage in each meter each time using
NLPC
100 and NRDGS 100 and then recording the STDEV, MEAN, MAX, and MIN
values.
What I have observed is that the Cal Constant is acceptably low but the
MEAN value per measurement is drifting up in one meter and drifting down
in the other. The unit drifting down has a new A3 board installed and
the
unit drifting upward is an Agilent 3458A only a few years old so would
not
be expected to have a drifting A3 board. It was only calibrated 1 time
per
the internal REV number, so would have likely been re-calibrated if the
A3
board was replaced.
My question is what is the likely cause of the drift in the MEAN voltage
reading if the Cal Constant value is relatively constant? SN 18 says the
drift rate of the Cal Constant is an indication of drift in the A3 AD
board, but I believe it assumes the A9 voltage ref board has a constant
value over time. If this is not true (e;g., the voltage ref is
drifting),
I think this would explain the drift in the voltage reading even though
the
Cal Constant is relatively constant. Any opinions on this?
Thanks,
Randy Evans
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
If you are confident that the 732A is stable to 0.1ppm then just run an acal afer a wek, and determnine what has driifted. Your A/D board, as can be seen with a drifty cal constant, or, els the A9 ref.
Dont do a checl everyday. this hides information.
Gesendet: Freitag, 01. Dezember 2017 um 19:49 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] Drifting 3458As
I am running the ACAL to check cal constant drift over several times a day
and over a 7 day period (per SN18). The varying times of the day are to
see how temperature sensitivity affects the cal constant drift since the
room temperature is not constant through the day but is relatively constant
at the same time of day each day; i.e., the furnace is off during the night
so mornings are about 5 degrees C cooler than the afternoon. Even after
ACAL, I am seeing up to an absolute difference of 3.5 uV on the 732A
voltage reading between the HP and Agilent 3458As, depending on the time of
day. I suspect the A9 board in the Agilent is the reason since the HP
3458A reads much closer to the 732A value than the Agilent 3458A over
temperature. Based on long term comparisons with other 732As and a 732B, I
am confident the 732A is stable within 0.1uV.
I am using an ebay A9 board in the Agilent 3458A for the current testing
since I suspect the original A9 board in the Agilent 3458A. I am
separately testing the original A9 board since I am not completely
confident in the ebay A9 board. I do not have enough data to draw a
conclusion yet about the original A9 board. My current concern is to see
if there is agreement that the A9 board is likely the culprit.
Thanks,
Randy Evans
On Fri, Dec 1, 2017 at 1:58 AM, acbern@gmx.de wrote:
What you describe (A9 drift) would be the explanation. However question is
how much you see. The A9 should not drift much in the intervals you talk
about (days).
Also, you should run the ACAL to determine the cal constant not several
times a day but with several days inbetween, and then divide by number of
days to determin drift
Frequent ACAL may not give you good results (random fluctuations such as
noise, temp... and their impact during cal gets higher the shorter the time
between ACALs)
Gesendet: Donnerstag, 30. November 2017 um 07:40 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] Drifting 3458As
I have been testing two 3458As against a known good Fluke 732A. Each
3458A
was calibrated for DCV before the start of the measurements so they both
started at the same point. I have been running a series of tests
consisting of measuring the Cal Constant as detailed in Service Note 18
for
each meter several times a day (to calculate the drift per the procedure
in
SN 18). I also measure the 732A voltage in each meter each time using
NLPC
100 and NRDGS 100 and then recording the STDEV, MEAN, MAX, and MIN
values.
What I have observed is that the Cal Constant is acceptably low but the
MEAN value per measurement is drifting up in one meter and drifting down
in the other. The unit drifting down has a new A3 board installed and
the
unit drifting upward is an Agilent 3458A only a few years old so would
not
be expected to have a drifting A3 board. It was only calibrated 1 time
per
the internal REV number, so would have likely been re-calibrated if the
A3
board was replaced.
My question is what is the likely cause of the drift in the MEAN voltage
reading if the Cal Constant value is relatively constant? SN 18 says the
drift rate of the Cal Constant is an indication of drift in the A3 AD
board, but I believe it assumes the A9 voltage ref board has a constant
value over time. If this is not true (e;g., the voltage ref is
drifting),
I think this would explain the drift in the voltage reading even though
the
Cal Constant is relatively constant. Any opinions on this?
Thanks,
Randy Evans
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.