MM
Matus Michael
Thu, Jun 30, 2016 12:51 PM
I have an old HP 5071 Cs-standard here on my desk. The oven of the (second) tube has lost its Cs completely so the instrument is no longer working. Since the electronic is quite old we acquired a complete new 5071A (now from Symmetricom).
I tried to "reactivate" the HP as a fancy desktop clock without success. Tried different commands on the serial bus but was not able to switch off the fatal error states (which is ok for the proper use). It is not possible to set Time and Date, nor perform a synchronisation. The display just stays dark. Also the reference oscillator is driven to one end causing a horrible frequency stability.
Does somebody know a procedure to override the error states so that the display works with a depleted CBT by programming? I do not want to modify the hardware. And to state it explicitly: I do not need a actual time/frequency standard just a mockup for my desk.
Michael
Dr. Michael Matus
BEV - Bundesamt fuer Eich- und Vermessungswesen
Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
(Head of Section, Dimensional Quantities, Frequency, Time)
Tel. +43 1 21110-6540, Fax +43 1 21110-996000
1160 Wien, Arltgasse 35, Austria
michael.matus@bev.gv.at
www.bev.gv.at
I have an old HP 5071 Cs-standard here on my desk. The oven of the (second) tube has lost its Cs completely so the instrument is no longer working. Since the electronic is quite old we acquired a complete new 5071A (now from Symmetricom).
I tried to "reactivate" the HP as a fancy desktop clock without success. Tried different commands on the serial bus but was not able to switch off the fatal error states (which is ok for the proper use). It is not possible to set Time and Date, nor perform a synchronisation. The display just stays dark. Also the reference oscillator is driven to one end causing a horrible frequency stability.
Does somebody know a procedure to override the error states so that the display works with a depleted CBT by programming? I do not want to modify the hardware. And to state it explicitly: I do not need a actual time/frequency standard just a mockup for my desk.
Michael
Dr. Michael Matus
-------------------------------------------------------------
BEV - Bundesamt fuer Eich- und Vermessungswesen
Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
(Head of Section, Dimensional Quantities, Frequency, Time)
Tel. +43 1 21110-6540, Fax +43 1 21110-996000
1160 Wien, Arltgasse 35, Austria
michael.matus@bev.gv.at
www.bev.gv.at
-------------------------------------------------------------
TV
Tom Van Baak
Thu, Jun 30, 2016 2:06 PM
From the front panel: Top -> Clock -> Set -> Display:ON
The same menu page also allows you to set the HH:MM:SS time.
This can be done within a few seconds of power up, regardless if the CBT is warming up, dead, or alive.
The amber attention LED will remain lit, but a dead CBT should not affect the red LED time display.
/tvb
----- Original Message -----
From: "Matus Michael" Michael.Matus@bev.gv.at
To: time-nuts@febo.com
Sent: Thursday, June 30, 2016 5:51 AM
Subject: [time-nuts] HP 5071A second life
I have an old HP 5071 Cs-standard here on my desk. The oven of the (second) tube has lost its Cs completely so the instrument is no longer working. Since the electronic is quite old we acquired a complete new 5071A (now from Symmetricom).
I tried to "reactivate" the HP as a fancy desktop clock without success. Tried different commands on the serial bus but was not able to switch off the fatal error states (which is ok for the proper use). It is not possible to set Time and Date, nor perform a synchronisation. The display just stays dark. Also the reference oscillator is driven to one end causing a horrible frequency stability.
Does somebody know a procedure to override the error states so that the display works with a depleted CBT by programming? I do not want to modify the hardware. And to state it explicitly: I do not need a actual time/frequency standard just a mockup for my desk.
Michael
Dr. Michael Matus
BEV - Bundesamt fuer Eich- und Vermessungswesen
Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
(Head of Section, Dimensional Quantities, Frequency, Time)
Tel. +43 1 21110-6540, Fax +43 1 21110-996000
1160 Wien, Arltgasse 35, Austria
michael.matus@bev.gv.at
www.bev.gv.at
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
>From the front panel: Top -> Clock -> Set -> Display:ON
The same menu page also allows you to set the HH:MM:SS time.
This can be done within a few seconds of power up, regardless if the CBT is warming up, dead, or alive.
The amber attention LED will remain lit, but a dead CBT should not affect the red LED time display.
/tvb
----- Original Message -----
From: "Matus Michael" <Michael.Matus@bev.gv.at>
To: <time-nuts@febo.com>
Sent: Thursday, June 30, 2016 5:51 AM
Subject: [time-nuts] HP 5071A second life
>I have an old HP 5071 Cs-standard here on my desk. The oven of the (second) tube has lost its Cs completely so the instrument is no longer working. Since the electronic is quite old we acquired a complete new 5071A (now from Symmetricom).
>
> I tried to "reactivate" the HP as a fancy desktop clock without success. Tried different commands on the serial bus but was not able to switch off the fatal error states (which is ok for the proper use). It is not possible to set Time and Date, nor perform a synchronisation. The display just stays dark. Also the reference oscillator is driven to one end causing a horrible frequency stability.
>
> Does somebody know a procedure to override the error states so that the display works with a depleted CBT by programming? I do not want to modify the hardware. And to state it explicitly: I do not need a actual time/frequency standard just a mockup for my desk.
>
> Michael
>
> Dr. Michael Matus
> -------------------------------------------------------------
> BEV - Bundesamt fuer Eich- und Vermessungswesen
> Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
> (Head of Section, Dimensional Quantities, Frequency, Time)
> Tel. +43 1 21110-6540, Fax +43 1 21110-996000
> 1160 Wien, Arltgasse 35, Austria
> michael.matus@bev.gv.at
> www.bev.gv.at
> -------------------------------------------------------------
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
PS
paul swed
Thu, Jun 30, 2016 2:07 PM
Michael
Thats going to be quite the clock. Very unfortunate but thats what happens
the Cs dry up. I seriously messed with a much older CS and was able to get
a few more C's out of the tube. The 5071 is smart so not sure any of what I
did makes any sense. Basically heating the oven to a slightly higher
temperature.
The other point that may not be of interest is that even dead 5071s seem to
have a great deal of value.
Regards
Paul
WB8TSL
On Thu, Jun 30, 2016 at 8:51 AM, Matus Michael Michael.Matus@bev.gv.at
wrote:
I have an old HP 5071 Cs-standard here on my desk. The oven of the
(second) tube has lost its Cs completely so the instrument is no longer
working. Since the electronic is quite old we acquired a complete new 5071A
(now from Symmetricom).
I tried to "reactivate" the HP as a fancy desktop clock without success.
Tried different commands on the serial bus but was not able to switch off
the fatal error states (which is ok for the proper use). It is not possible
to set Time and Date, nor perform a synchronisation. The display just stays
dark. Also the reference oscillator is driven to one end causing a horrible
frequency stability.
Does somebody know a procedure to override the error states so that the
display works with a depleted CBT by programming? I do not want to modify
the hardware. And to state it explicitly: I do not need a actual
time/frequency standard just a mockup for my desk.
Michael
Dr. Michael Matus
BEV - Bundesamt fuer Eich- und Vermessungswesen
Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
(Head of Section, Dimensional Quantities, Frequency, Time)
Tel. +43 1 21110-6540, Fax +43 1 21110-996000
1160 Wien, Arltgasse 35, Austria
michael.matus@bev.gv.at
www.bev.gv.at
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Michael
Thats going to be quite the clock. Very unfortunate but thats what happens
the Cs dry up. I seriously messed with a much older CS and was able to get
a few more C's out of the tube. The 5071 is smart so not sure any of what I
did makes any sense. Basically heating the oven to a slightly higher
temperature.
The other point that may not be of interest is that even dead 5071s seem to
have a great deal of value.
Regards
Paul
WB8TSL
On Thu, Jun 30, 2016 at 8:51 AM, Matus Michael <Michael.Matus@bev.gv.at>
wrote:
> I have an old HP 5071 Cs-standard here on my desk. The oven of the
> (second) tube has lost its Cs completely so the instrument is no longer
> working. Since the electronic is quite old we acquired a complete new 5071A
> (now from Symmetricom).
>
> I tried to "reactivate" the HP as a fancy desktop clock without success.
> Tried different commands on the serial bus but was not able to switch off
> the fatal error states (which is ok for the proper use). It is not possible
> to set Time and Date, nor perform a synchronisation. The display just stays
> dark. Also the reference oscillator is driven to one end causing a horrible
> frequency stability.
>
> Does somebody know a procedure to override the error states so that the
> display works with a depleted CBT by programming? I do not want to modify
> the hardware. And to state it explicitly: I do not need a actual
> time/frequency standard just a mockup for my desk.
>
> Michael
>
> Dr. Michael Matus
> -------------------------------------------------------------
> BEV - Bundesamt fuer Eich- und Vermessungswesen
> Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
> (Head of Section, Dimensional Quantities, Frequency, Time)
> Tel. +43 1 21110-6540, Fax +43 1 21110-996000
> 1160 Wien, Arltgasse 35, Austria
> michael.matus@bev.gv.at
> www.bev.gv.at
> -------------------------------------------------------------
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
PT
Peter Torry
Thu, Jun 30, 2016 2:14 PM
I have been given a 10MHz oscillator that appears to be in good physical
condition but I have no further information on it hence this request.
It is marked as follows:- Quartzkeramic 2006/10SCD-531
10.000MHz 12v=
K 837383
Any information such as connection details, performance etc. will be
greatly appreciated.
Regards
Peter
I have been given a 10MHz oscillator that appears to be in good physical
condition but I have no further information on it hence this request.
It is marked as follows:- Quartzkeramic 2006/10SCD-531
10.000MHz 12v=
K 837383
Any information such as connection details, performance etc. will be
greatly appreciated.
Regards
Peter
MM
Matus Michael
Fri, Jul 1, 2016 9:26 AM
Thank you very much. I was too stupid to try the obvious way!
Michael
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf
Of Tom Van Baak
Sent: Thursday, June 30, 2016 4:06 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 5071A second life
From the front panel: Top -> Clock -> Set -> Display:ON
The same menu page also allows you to set the HH:MM:SS time.
This can be done within a few seconds of power up, regardless
if the CBT is warming up, dead, or alive.
The amber attention LED will remain lit, but a dead CBT
should not affect the red LED time display.
/tvb
----- Original Message -----
From: "Matus Michael" Michael.Matus@bev.gv.at
To: time-nuts@febo.com
Sent: Thursday, June 30, 2016 5:51 AM
Subject: [time-nuts] HP 5071A second life
I have an old HP 5071 Cs-standard here on my desk. The oven
of the (second) tube has lost its Cs completely so the
instrument is no longer working. Since the electronic is
quite old we acquired a complete new 5071A (now from Symmetricom).
I tried to "reactivate" the HP as a fancy desktop clock
without success. Tried different commands on the serial bus
but was not able to switch off the fatal error states (which
is ok for the proper use). It is not possible to set Time and
Date, nor perform a synchronisation. The display just stays
dark. Also the reference oscillator is driven to one end
causing a horrible frequency stability.
Does somebody know a procedure to override the error states
so that the display works with a depleted CBT by programming?
I do not want to modify the hardware. And to state it
explicitly: I do not need a actual time/frequency standard
just a mockup for my desk.
Michael
Dr. Michael Matus
BEV - Bundesamt fuer Eich- und Vermessungswesen
Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
(Head of Section, Dimensional Quantities, Frequency, Time)
Tel. +43 1 21110-6540, Fax +43 1 21110-996000
1160 Wien, Arltgasse 35, Austria
michael.matus@bev.gv.at
www.bev.gv.at
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
and follow the instructions there.
Thank you very much. I was too stupid to try the obvious way!
Michael
> -----Original Message-----
> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf
> Of Tom Van Baak
> Sent: Thursday, June 30, 2016 4:06 PM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] HP 5071A second life
>
> From the front panel: Top -> Clock -> Set -> Display:ON
> The same menu page also allows you to set the HH:MM:SS time.
> This can be done within a few seconds of power up, regardless
> if the CBT is warming up, dead, or alive.
> The amber attention LED will remain lit, but a dead CBT
> should not affect the red LED time display.
>
> /tvb
>
> ----- Original Message -----
> From: "Matus Michael" <Michael.Matus@bev.gv.at>
> To: <time-nuts@febo.com>
> Sent: Thursday, June 30, 2016 5:51 AM
> Subject: [time-nuts] HP 5071A second life
>
>
> >I have an old HP 5071 Cs-standard here on my desk. The oven
> of the (second) tube has lost its Cs completely so the
> instrument is no longer working. Since the electronic is
> quite old we acquired a complete new 5071A (now from Symmetricom).
> >
> > I tried to "reactivate" the HP as a fancy desktop clock
> without success. Tried different commands on the serial bus
> but was not able to switch off the fatal error states (which
> is ok for the proper use). It is not possible to set Time and
> Date, nor perform a synchronisation. The display just stays
> dark. Also the reference oscillator is driven to one end
> causing a horrible frequency stability.
> >
> > Does somebody know a procedure to override the error states
> so that the display works with a depleted CBT by programming?
> I do not want to modify the hardware. And to state it
> explicitly: I do not need a actual time/frequency standard
> just a mockup for my desk.
> >
> > Michael
> >
> > Dr. Michael Matus
> > -------------------------------------------------------------
> > BEV - Bundesamt fuer Eich- und Vermessungswesen
> > Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
> > (Head of Section, Dimensional Quantities, Frequency, Time)
> > Tel. +43 1 21110-6540, Fax +43 1 21110-996000
> > 1160 Wien, Arltgasse 35, Austria
> > michael.matus@bev.gv.at
> > www.bev.gv.at
> > -------------------------------------------------------------
> > _______________________________________________
> > time-nuts mailing list -- time-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>