Hi, all
I've done some experiments on my 53230A, and I've come across an issue I
think could be worth knowing about.
I have not tested this on other 53230A's - if anyone has a 53230A and some
spare time, I'd be grateful if someone could try to replicate.
In a nutshell, the first frequency measurement following an INIT is biased.
The magnitude of the bias is dependant on measurement mode (CONT/RCON), and
gate time.
It is worth noting that every READ implicitly calls INIT - so any software
collecting a time series using repeated calls to READ to fetch single
readings will give biased results. Not for the first sample in the series,
but all of them.
As an example, the last time series I collected this way using RCON and
0.1s gatetime gave an average frequency error of -5.2e-10.
A writeup is here: http://www.efos3.com/HPAK53230A-1.html
Best regards,
Ole
Ole,
Is this part of the behavior that the Norwegian NMI was presenting at a
poster session at EFTF 2016 in York?
I remember they borrowed your 53230A for that work.
Cheers,
Magnus
On 11/23/2017 09:05 PM, Ole Petter Ronningen wrote:
Hi, all
I've done some experiments on my 53230A, and I've come across an issue I
think could be worth knowing about.
I have not tested this on other 53230A's - if anyone has a 53230A and some
spare time, I'd be grateful if someone could try to replicate.
In a nutshell, the first frequency measurement following an INIT is biased.
The magnitude of the bias is dependant on measurement mode (CONT/RCON), and
gate time.
It is worth noting that every READ implicitly calls INIT - so any software
collecting a time series using repeated calls to READ to fetch single
readings will give biased results. Not for the first sample in the series,
but all of them.
As an example, the last time series I collected this way using RCON and
0.1s gatetime gave an average frequency error of -5.2e-10.
A writeup is here: http://www.efos3.com/HPAK53230A-1.html
Best regards,
Ole
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi Magnus
No, I believe this is a separate issue. The data for that paper was
collected gap free, and was not affected by this particular issue.
It may be related somehow - as presented in that paper, data collected in
CONT mode shows a slope of tau^-1/2 on ADEV.
Data collected gap free in RCON-mode (i.e. not by repeated calls to READ)
shows the expected slope of Tau^-1.
The issue I point out here is that if data is collected using repeated
calls to "READ", even data collected in RCON-mode shows a slope of
tau^-1/2, in addition to a frequency bias. The magnitude of the frequency
bias is gate time dependent.
Ole
On Fri, Nov 24, 2017 at 2:45 PM, Magnus Danielson <
magnus@rubidium.dyndns.org> wrote:
Ole,
Is this part of the behavior that the Norwegian NMI was presenting at a
poster session at EFTF 2016 in York?
I remember they borrowed your 53230A for that work.
Cheers,
Magnus
On 11/23/2017 09:05 PM, Ole Petter Ronningen wrote:
Hi, all
I've done some experiments on my 53230A, and I've come across an issue I
think could be worth knowing about.
I have not tested this on other 53230A's - if anyone has a 53230A and some
spare time, I'd be grateful if someone could try to replicate.
In a nutshell, the first frequency measurement following an INIT is
biased.
The magnitude of the bias is dependant on measurement mode (CONT/RCON),
and
gate time.
It is worth noting that every READ implicitly calls INIT - so any software
collecting a time series using repeated calls to READ to fetch single
readings will give biased results. Not for the first sample in the series,
but all of them.
As an example, the last time series I collected this way using RCON and
0.1s gatetime gave an average frequency error of -5.2e-10.
A writeup is here: http://www.efos3.com/HPAK53230A-1.html
Best regards,
Ole
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.
Although, as I read the paper again, it is possible that the frequency bias
in the 10 second CONT-mode measurements could come from this effect.
That particular series (and only that) was collected using TimeLab. I
supplied them with a custom compiled driver that does not use repeated
calls to READ - and therefore does not exhibit this particular bias. I
always assumed that this driver was used to collect that series, but the
paper does not state this explicitly so I might be wrong.
The bias I am pointing to is gate time dependant, the bias getting smaller
with longer gate times. It is possible it will be at the E-15 level at 10
second gatetime, like in the paper.
Ole
On Fri, Nov 24, 2017 at 2:45 PM, Magnus Danielson <
magnus@rubidium.dyndns.org> wrote:
Ole,
Is this part of the behavior that the Norwegian NMI was presenting at a
poster session at EFTF 2016 in York?
I remember they borrowed your 53230A for that work.
Cheers,
Magnus
On 11/23/2017 09:05 PM, Ole Petter Ronningen wrote:
Hi, all
I've done some experiments on my 53230A, and I've come across an issue I
think could be worth knowing about.
I have not tested this on other 53230A's - if anyone has a 53230A and some
spare time, I'd be grateful if someone could try to replicate.
In a nutshell, the first frequency measurement following an INIT is
biased.
The magnitude of the bias is dependant on measurement mode (CONT/RCON),
and
gate time.
It is worth noting that every READ implicitly calls INIT - so any software
collecting a time series using repeated calls to READ to fetch single
readings will give biased results. Not for the first sample in the series,
but all of them.
As an example, the last time series I collected this way using RCON and
0.1s gatetime gave an average frequency error of -5.2e-10.
A writeup is here: http://www.efos3.com/HPAK53230A-1.html
Best regards,
Ole