I pickup up an HP/Agilent (front says HP, rear says Agilent) 4339B
high-resistance meter from eBay. The 4339B has an internal supply up to
1000 V, an ammeter to measure current, and obviously computes resistance
from Ohms Law. It is supposed to work up to 16 Peta ohms (1.6x10^16 ohms),
which would pass a current of 62.5 fA with 1000 V.
It was sold as non-working, with no output. Even attempting to get some
output was challenging as it had an obscure 6-pin interlock connector, and
there's no information in the manual about this. It also has a trixaxial
input connector with a screw thread. I asked Keysight for some information
about the connectors. I still don't know what the triaxial connector is,
but I have the interlock sussed - a HiroseHR10A-7P-6P(73) plug fits, and
there's some information about the wiring of this in the manual for the
11617C low-noise test leads.
Keysight offered to do a free technical evaluation, so I thought I'd take
them up on it, rather than waste too much time trying myself to get it
working. It was sent to Keysight and they sent me a report showing there is
output voltage, despite it being sold as having no output, and the output
is all in spec. Unfortunately, it was out of spec on one of the current
ranges. It is reading 0.082 nA low with a test current of 10 nA, but the
specification is +/- 0.063 nA. So it is only a little bit out, but it is
still out of spec. Test currents of 100 pA, 1 nA, 10 nA, 100 nA, 1 uA, 10
uA and 100 uA were used, so the 10 nA is neither at the low nor high end of
the range of the meter.
When I received an email copy of the free technical report, I assumed the
4339B had not been subjected to a full calibration as it was only tested
with resistors up to 10^11 ohms, but it is supposed to read resistance up
to 1.6 x 10^16 ohm, so the highest value resistor used during the technical
evualation is 5 orders of magnitude lower than the what the meter can read.
Also the meter was only tested at 100 V when reading resistors, whereas I
would have thought it more sensible to test at a higher voltage, given it
can output 1 kV.
I decided to pay Keysight for a full calibration, so I got a cal
certificate, which shows the out of spec reading.
http://www.kirkbymicrowave.co.uk/tmp/1-9690444179-1-combined-file.pdf
I was rather hoping the cal certificate would have some extra readings, but
I actually received no more technical information than Keysight had given
me for nothing, although they did update the firmware as I requested.
I have two questions.
Is 10^11 ohms a practical limit on making resistors that can be used for
calibration purposes? I don't know what the Keysight uncertainty on their
calibration standards are, so maybe 10^11 ohms is all its practical to
make.
Does anyone know if the meter can be adjusted on individual ranges? I
find it a bit hard to believe that given the ammeter was tested with
currents covering 6 orders of magnitude (100 pA to 100 uA), that there's no
adjustment on individual ranges. I doubt that it is possible to design a
circuit to cover that range, without switching in some different resistors
at some point.
But I'm told it can't be adjusted, and needs a new board.
I'm wondering if anyone knows anything about this meter, just in case
Keysight are wrong. I know they should not be, but it is not a current
instrument, so it's just possible the engineer did not know it could be
adjusted.
Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford,
Essex, CM3 6DT, United Kingdom
Registered in England and Wales as company number 08914892
http://www.kirkbymicrowave.co.uk/
Tel 01621'680100 / +44 1621-680100
Maybe one bad resistor? I've read that tin whiskering can occur even on relatively recent equipment, slower than lead free but still a problem.
I'd do a visual inspection just to see, shouldn't be an issue.
-A
From: volt-nuts volt-nuts-bounces@febo.com on behalf of Dr. David Kirkby drkirkby@kirkbymicrowave.co.uk
Sent: 02 March 2018 03:16:24
To: Discussion of precise voltage measurement
Subject: [volt-nuts] Precision high resistance measurements / calibration of HP 4339B high-resistance meter.
I pickup up an HP/Agilent (front says HP, rear says Agilent) 4339B
high-resistance meter from eBay. The 4339B has an internal supply up to
1000 V, an ammeter to measure current, and obviously computes resistance
from Ohms Law. It is supposed to work up to 16 Peta ohms (1.6x10^16 ohms),
which would pass a current of 62.5 fA with 1000 V.
It was sold as non-working, with no output. Even attempting to get some
output was challenging as it had an obscure 6-pin interlock connector, and
there's no information in the manual about this. It also has a trixaxial
input connector with a screw thread. I asked Keysight for some information
about the connectors. I still don't know what the triaxial connector is,
but I have the interlock sussed - a HiroseHR10A-7P-6P(73) plug fits, and
there's some information about the wiring of this in the manual for the
11617C low-noise test leads.
Keysight offered to do a free technical evaluation, so I thought I'd take
them up on it, rather than waste too much time trying myself to get it
working. It was sent to Keysight and they sent me a report showing there is
output voltage, despite it being sold as having no output, and the output
is all in spec. Unfortunately, it was out of spec on one of the current
ranges. It is reading 0.082 nA low with a test current of 10 nA, but the
specification is +/- 0.063 nA. So it is only a little bit out, but it is
still out of spec. Test currents of 100 pA, 1 nA, 10 nA, 100 nA, 1 uA, 10
uA and 100 uA were used, so the 10 nA is neither at the low nor high end of
the range of the meter.
When I received an email copy of the free technical report, I assumed the
4339B had not been subjected to a full calibration as it was only tested
with resistors up to 10^11 ohms, but it is supposed to read resistance up
to 1.6 x 10^16 ohm, so the highest value resistor used during the technical
evualation is 5 orders of magnitude lower than the what the meter can read.
Also the meter was only tested at 100 V when reading resistors, whereas I
would have thought it more sensible to test at a higher voltage, given it
can output 1 kV.
I decided to pay Keysight for a full calibration, so I got a cal
certificate, which shows the out of spec reading.
http://www.kirkbymicrowave.co.uk/tmp/1-9690444179-1-combined-file.pdf
I was rather hoping the cal certificate would have some extra readings, but
I actually received no more technical information than Keysight had given
me for nothing, although they did update the firmware as I requested.
I have two questions.
Is 10^11 ohms a practical limit on making resistors that can be used for
calibration purposes? I don't know what the Keysight uncertainty on their
calibration standards are, so maybe 10^11 ohms is all its practical to
make.
Does anyone know if the meter can be adjusted on individual ranges? I
find it a bit hard to believe that given the ammeter was tested with
currents covering 6 orders of magnitude (100 pA to 100 uA), that there's no
adjustment on individual ranges. I doubt that it is possible to design a
circuit to cover that range, without switching in some different resistors
at some point.
But I'm told it can't be adjusted, and needs a new board.
I'm wondering if anyone knows anything about this meter, just in case
Keysight are wrong. I know they should not be, but it is not a current
instrument, so it's just possible the engineer did not know it could be
adjusted.
Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford,
Essex, CM3 6DT, United Kingdom
Registered in England and Wales as company number 08914892
http://www.kirkbymicrowave.co.uk/
Tel 01621'680100 / +44 1621-680100
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
On 2 March 2018 at 09:22, Andre Andre@lanoe.net wrote:
Maybe one bad resistor? I've read that tin whiskering can occur even on
relatively recent equipment, slower than lead free but still a problem.
I'd do a visual inspection just to see, shouldn't be an issue.
-A
An issue with fingerprints would not totally surprise me. It is obvious the
seller had multiple meters, which were sent to a lab, which I am suspicious
of*. The seller said the lab would have swapped pieces between units to try
to get one working. He had at least 3.
My one, which has a recent(ish) Agilent serial number, and Agilent
written on the back, but an HP front panel.
Another which has HP on the back, but an Agilent front panel.
Another said to be working.
So its almost certain bits have been swapped between different units.
It is is finger prints, what would be the best way to remove them? I see
some say IPA, but others have said that can disolve solder resist.
I once had a Sun workstation which had one of the covers missing of the
back where sbus expansion cards
https://en.wikipedia.org/wiki/SBus
were supposed to be put. This was stored in the garage. Some mice decided
this Sun would make a nice home, and used it as a toilet! Needles to say
the board was covered in what mice excrete. After various attempts to get
it back to life, an hour or so in the bath with plain tap water did work.
But of course very low currents of the 4339B would prevent different
challenges to a digital board, where signals are either 0 or 1. But whilst
the 4339B may not be free of finger prints, it has not been used as a mouse
loo. So if it does require cleaning, it should be fairly modest, but I have
no idea what to use.
I buy a IPA on eBay for cleaning, which is claimed to be 99.9% pure, but I
doubt it is really that pure, so if IPA was used, I would buy from a proper
chemical supplier.
Dave
Oops - I think I didn't send this message properly yesterday - here goes
again. Ed
Yes, David, unless you go to very extreme measures, you won't see real R
values that have any practical meaning beyond E12 ohms or so. Most
practical insulation Rs may be around E12-E14 tops, unless you go to
sapphire. Up in that region, the R may be all within a material, or
include surface components like a film of dirt or moisture, or a
fingerprint.
E11 resistors can be made to fairly high precision, and maybe E12
nowadays. In the old days, higher values were made by stacking E11s -
like ten in series to get E12 with decent precision. The glass packaging
also limits how high it can go, due to leakage within and on the
surface. I once used a glass reed relay capsule as an ultra-high
resistance in a circuit. There was no precision or stability at all, but
it made a nice high resistor (probably E14-ish dry) even though there
was no element in there, and the circuit didn't care, as long as it was
very high, but not infinite.
The specs on this HP unit are likely just the most extreme capability
taking maximum voltage over minimum current resolution, but any
measurements would tend to be very noisy and unstable anyway. Also,
testing at the extreme 1 kV makes the numbers seem more impressive, but
the voltage coefficient of resistance will pretty much be unpredictable.
If this is a digital meter, then the other spec trick that tends to
obscure the real performance limit is that the ultimate resolution and
noise is that last digit - or even last two or three - that may may be
pretty jumpy, unless very long averaging time is used.
There may be newer, fancier electrometers nowadays, but Keithley used to
be the standard for these in the old days, before several digits of DVM
resolution complicated the specs. They had a vibrating capacitor
electrometer with all-sapphire input structure back in the 1970s/80s I
think, that was the epitome of electrometers. I forget the model number,
but vaguely recall that it could reach the aA region full scale - not
that last digit of resolution thing. It's long obsolete, and I don't
think they ever made anything actually better - only added DVM digits to
less capable, conventional semiconductor amplifier techniques. If you
can find info on it, it's an interesting read. I found a pdf of the
manual years ago, but have no idea where it is now, or what info may
still be around.
Ed
Here are specs from an older General Radio Bridge for observation.
GR 1644-A specifications:
Resistance Range: 1 kΩ to 1000 TΩ (10^3 to 10^15 Ω) in ten ranges.
Accuracy: 10^3 Ω to 10^10 Ω, ±1 %. After self-calibration: 10^10 to
10^12 Ω, ±1%*; 10^13 Ω, ±2%; 10^14 Ω, ±10%; 10^15 Ω, ± one scale
division.
ΔR% Dial: ±5% range; accurate to ±0.2% or, for small changes,
to ±0.1%.
Test Voltage: Voltage accuracy is ±3% ±0.5 V.
Fixed Voltages** 10 20 50 100 200 500 1000 v
Best regards,
mitch
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of ed breya
Sent: Friday, March 02, 2018 12:23 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Precision high resistance measurements / calibration of HP 4339B high-resistance meter.
Oops - I think I didn't send this message properly yesterday - here goes again. Ed
Yes, David, unless you go to very extreme measures, you won't see real R values that have any practical meaning beyond E12 ohms or so. Most practical insulation Rs may be around E12-E14 tops, unless you go to sapphire. Up in that region, the R may be all within a material, or include surface components like a film of dirt or moisture, or a fingerprint.
E11 resistors can be made to fairly high precision, and maybe E12 nowadays. In the old days, higher values were made by stacking E11s - like ten in series to get E12 with decent precision. The glass packaging also limits how high it can go, due to leakage within and on the surface. I once used a glass reed relay capsule as an ultra-high resistance in a circuit. There was no precision or stability at all, but it made a nice high resistor (probably E14-ish dry) even though there was no element in there, and the circuit didn't care, as long as it was very high, but not infinite.
The specs on this HP unit are likely just the most extreme capability taking maximum voltage over minimum current resolution, but any measurements would tend to be very noisy and unstable anyway. Also, testing at the extreme 1 kV makes the numbers seem more impressive, but the voltage coefficient of resistance will pretty much be unpredictable.
If this is a digital meter, then the other spec trick that tends to obscure the real performance limit is that the ultimate resolution and noise is that last digit - or even last two or three - that may may be pretty jumpy, unless very long averaging time is used.
There may be newer, fancier electrometers nowadays, but Keithley used to be the standard for these in the old days, before several digits of DVM resolution complicated the specs. They had a vibrating capacitor electrometer with all-sapphire input structure back in the 1970s/80s I think, that was the epitome of electrometers. I forget the model number, but vaguely recall that it could reach the aA region full scale - not that last digit of resolution thing. It's long obsolete, and I don't think they ever made anything actually better - only added DVM digits to less capable, conventional semiconductor amplifier techniques. If you can find info on it, it's an interesting read. I found a pdf of the manual years ago, but have no idea where it is now, or what info may still be around.
Ed
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.