I'm not sure what the Z3801A uses for EFC range. Heather uses the EFC relative command to report the EFC setting. That command reports values from -100% to 100%. There is a command that reports the DAC input in counts, but nothing documented that shows volts/count.
Since the DAC is at -2V with a 99+% DAC setting, it looks like the DAC circuit has issues, or power supply problems. I think the DAC circuit is a 16 bit DAC with 4 more bits of resolution created by dithering the DAC inputs.
Hi
The control range should be +/- 5V if I remember correctly. If it’s not, the
only other standard would be 0-5V. There is zero logic in running a wider
swing mod to an OCXO at a lower EFC range.
It indeed sounds like the DAC has trouble.
Bob
On Mar 6, 2018, at 1:24 AM, Mark Sims holrum@hotmail.com wrote:
I'm not sure what the Z3801A uses for EFC range. Heather uses the EFC relative command to report the EFC setting. That command reports values from -100% to 100%. There is a command that reports the DAC input in counts, but nothing documented that shows volts/count.
Since the DAC is at -2V with a 99+% DAC setting, it looks like the DAC circuit has issues, or power supply problems. I think the DAC circuit is a 16 bit DAC with 4 more bits of resolution created by dithering the DAC inputs.
time-nuts mailing list -- time-nuts@febo.com