@ Poul-Henning:
It's improbable that we'll see improved specs.
The presentation of the Black Edition already happened end of May, and
the validation obviously was already running.
The description @ meilhaus already claims that no change in spec. will
happen, and the linked Black Edition specs are really identical.
@ David and Bob:
RoHS is a European requirement only, and may not (yet) apply to other
markets.
KS obviously has stocked a bunch of lead containing units (they called
it EU3458A) in their own premises before the exemption expired, and as
this stock would be depleted later, they needed to design a lead free
unit - initially for the EU market only.
I suppose they will sell the white version outside the EU for several
years, until their old component stock is depleted also.
I wonder what will happen to the EU34420A, which has the same problem,
but lead free soldering might pose greater problems, due to e.m.f. @ nV
level.
@ Tom:
I can't tell anything about Golden Units, as I only own a regular, white
HP3458A from 2001, which still performs better than specification..
+1.8ppm drift only, over 9 years for DCV, and Ohm is also much more
stable than specification, as far as I can judge.
Ok, the trick is, that it does not run 24/365, and my LTZ reference
board is set to ~ 65°C.
Concerning the internal A/D converter clock, there are several papers
about AC metrology, especially on AC power measurements, where the PTB
improved the accuracy greatly by optically linking the internal clock to
the rear of the 3458A, to mitigate the 10 or 100ns (?) clock phase
jitter. See also Rado Lapuh, Sampling With 3458A, page 149.
A practical schematic is not published, so it's very appreciated that
you provide one, but a description, how to apply this clock would be
necessary.
Frank
Hi,
one should always keep in mind that all the specifications are made to
achieve a certain yield in production. Assuming a Gaussian distribution it
is not very unlikely, that we all have 3458As which perform much better
than specified. In that respect you cannot rely on vishay marketing.
Reducing the LTZ1000 temperature is also not that easy. The 3458A might be
a metrology grade DMM, but it is also a system DMM. Even on my bench the
TEMP? sometimes goes up to 47°C (at 26°C room) and it might be even higher
in the summer. And it is not in a rack and only a HP 3488A is sitting on
top which produces almost no heat. Running the 3458A in a rack full of
equipment will not allow to reduce the LTZ1000 that much.
I think the gap to the Fluke 8588A is too wide to catch up with only a
little bit improved specs. I think the main reason for the black edition is
to ensure that many test rigs can work for the next years.
Best regards
Philipp
Am Sa., 26. Okt. 2019 um 20:48 Uhr schrieb Frank Stellmach <
frank.stellmach@freenet.de>:
@ Poul-Henning:
It's improbable that we'll see improved specs.
The presentation of the Black Edition already happened end of May, and
the validation obviously was already running.
The description @ meilhaus already claims that no change in spec. will
happen, and the linked Black Edition specs are really identical.
@ David and Bob:
RoHS is a European requirement only, and may not (yet) apply to other
markets.
KS obviously has stocked a bunch of lead containing units (they called
it EU3458A) in their own premises before the exemption expired, and as
this stock would be depleted later, they needed to design a lead free
unit - initially for the EU market only.
I suppose they will sell the white version outside the EU for several
years, until their old component stock is depleted also.
I wonder what will happen to the EU34420A, which has the same problem,
but lead free soldering might pose greater problems, due to e.m.f. @ nV
level.
@ Tom:
I can't tell anything about Golden Units, as I only own a regular, white
HP3458A from 2001, which still performs better than specification..
+1.8ppm drift only, over 9 years for DCV, and Ohm is also much more
stable than specification, as far as I can judge.
Ok, the trick is, that it does not run 24/365, and my LTZ reference
board is set to ~ 65°C.
Concerning the internal A/D converter clock, there are several papers
about AC metrology, especially on AC power measurements, where the PTB
improved the accuracy greatly by optically linking the internal clock to
the rear of the 3458A, to mitigate the 10 or 100ns (?) clock phase
jitter. See also Rado Lapuh, Sampling With 3458A, page 149.
A practical schematic is not published, so it's very appreciated that
you provide one, but a description, how to apply this clock would be
necessary.
Frank
volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.
When I refer to a "Golden" 3458A is simple mean that over the years I have gone through dozens of 3458A's and in that time have found a few truly exceptional instruments that I and others who have them often refer to as "Golden". The biggest difference I see unit to unit is not ref/zenner but the A/D. A really good "Golden" 3458A can often stay at 5@10-8 (a few counts on the eighth digit) connected to a PJVS or good 732B for weeks or even months at a time. If you are lucky enough to have one don't ever let it go. So my thought was with advances since 1988 especially in A/D's will most/or a greater percentage of these redesigns be golden.
Cheers;
Tom Knox
303-554-0307
"Peace is not the absence of violence, but the presence of Justice" Both MLK and Albert Einstein
From: volt-nuts volt-nuts-bounces@lists.febo.com on behalf of Frank Stellmach frank.stellmach@freenet.de
Sent: Saturday, October 26, 2019 12:39 PM
To: volt-nuts@lists.febo.com volt-nuts@lists.febo.com
Subject: [volt-nuts] HP3458A - Black Edition
@ Poul-Henning:
It's improbable that we'll see improved specs.
The presentation of the Black Edition already happened end of May, and
the validation obviously was already running.
The description @ meilhaus already claims that no change in spec. will
happen, and the linked Black Edition specs are really identical.
@ David and Bob:
RoHS is a European requirement only, and may not (yet) apply to other
markets.
KS obviously has stocked a bunch of lead containing units (they called
it EU3458A) in their own premises before the exemption expired, and as
this stock would be depleted later, they needed to design a lead free
unit - initially for the EU market only.
I suppose they will sell the white version outside the EU for several
years, until their old component stock is depleted also.
I wonder what will happen to the EU34420A, which has the same problem,
but lead free soldering might pose greater problems, due to e.m.f. @ nV
level.
@ Tom:
I can't tell anything about Golden Units, as I only own a regular, white
HP3458A from 2001, which still performs better than specification..
+1.8ppm drift only, over 9 years for DCV, and Ohm is also much more
stable than specification, as far as I can judge.
Ok, the trick is, that it does not run 24/365, and my LTZ reference
board is set to ~ 65°C.
Concerning the internal A/D converter clock, there are several papers
about AC metrology, especially on AC power measurements, where the PTB
improved the accuracy greatly by optically linking the internal clock to
the rear of the 3458A, to mitigate the 10 or 100ns (?) clock phase
jitter. See also Rado Lapuh, Sampling With 3458A, page 149.
A practical schematic is not published, so it's very appreciated that
you provide one, but a description, how to apply this clock would be
necessary.
Frank
volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.
Sorry if too far off topic -Depending on the age of the 3458A if it uses the Dallas chips, SnapHat, or coin style battery. The older units had processor boards that filled the entire compartment and used Dallas chips. I always refitted those Dallas chips with sockets, but there is no shortcut to a data dump or full re-cal. The latest processor board (I have seen) uses a "coin" style battery. Loss of data or data dump during battery replacement can be a real problem. When dealing with stand along batteries I often "cheat" soldering in a couple temporary pins/eyelets to connect an external battery allowing me to remove the original battery without losing power/data during replacement. I then (if not already installed) add a battery socket with two permanent eyelets on +/- or the socket so replacement only takes a few minutes in the future.
Sadly this does not help with SnapHats and Dallas chips replacement. Although I wonder if a chip socket could be modified with +/- eyelet the stacks between the Snaphat battery and chip allowing easier future replacement.The point is I have so many pieces of equipment with critical data that I usually end up only replacing when the instrument dies. But once these up-grades are complete can save a lot of time allowing regular replacements.
Tom Knox
303-554-0307
"Peace is not the absence of violence, but the presence of Justice" Both MLK and Albert Einstein
From: volt-nuts volt-nuts-bounces@lists.febo.com on behalf of Philipp Cochems via volt-nuts volt-nuts@lists.febo.com
Sent: Saturday, October 26, 2019 1:13 PM
To: Discussion of precise voltage measurement volt-nuts@lists.febo.com
Cc: Philipp Cochems philipp.cochems@googlemail.com
Subject: Re: [volt-nuts] HP3458A - Black Edition
Hi,
one should always keep in mind that all the specifications are made to
achieve a certain yield in production. Assuming a Gaussian distribution it
is not very unlikely, that we all have 3458As which perform much better
than specified. In that respect you cannot rely on vishay marketing.
Reducing the LTZ1000 temperature is also not that easy. The 3458A might be
a metrology grade DMM, but it is also a system DMM. Even on my bench the
TEMP? sometimes goes up to 47°C (at 26°C room) and it might be even higher
in the summer. And it is not in a rack and only a HP 3488A is sitting on
top which produces almost no heat. Running the 3458A in a rack full of
equipment will not allow to reduce the LTZ1000 that much.
I think the gap to the Fluke 8588A is too wide to catch up with only a
little bit improved specs. I think the main reason for the black edition is
to ensure that many test rigs can work for the next years.
Best regards
Philipp
Am Sa., 26. Okt. 2019 um 20:48 Uhr schrieb Frank Stellmach <
frank.stellmach@freenet.de>:
@ Poul-Henning:
It's improbable that we'll see improved specs.
The presentation of the Black Edition already happened end of May, and
the validation obviously was already running.
The description @ meilhaus already claims that no change in spec. will
happen, and the linked Black Edition specs are really identical.
@ David and Bob:
RoHS is a European requirement only, and may not (yet) apply to other
markets.
KS obviously has stocked a bunch of lead containing units (they called
it EU3458A) in their own premises before the exemption expired, and as
this stock would be depleted later, they needed to design a lead free
unit - initially for the EU market only.
I suppose they will sell the white version outside the EU for several
years, until their old component stock is depleted also.
I wonder what will happen to the EU34420A, which has the same problem,
but lead free soldering might pose greater problems, due to e.m.f. @ nV
level.
@ Tom:
I can't tell anything about Golden Units, as I only own a regular, white
HP3458A from 2001, which still performs better than specification..
+1.8ppm drift only, over 9 years for DCV, and Ohm is also much more
stable than specification, as far as I can judge.
Ok, the trick is, that it does not run 24/365, and my LTZ reference
board is set to ~ 65°C.
Concerning the internal A/D converter clock, there are several papers
about AC metrology, especially on AC power measurements, where the PTB
improved the accuracy greatly by optically linking the internal clock to
the rear of the 3458A, to mitigate the 10 or 100ns (?) clock phase
jitter. See also Rado Lapuh, Sampling With 3458A, page 149.
A practical schematic is not published, so it's very appreciated that
you provide one, but a description, how to apply this clock would be
necessary.
Frank
volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.
Wow, +21C Delta on TEMP? is really huge for just a meter sitting alone on desk. There must be hot air blowing onto unit or dirty filter or faulty fan.
Even with old style fan cup, normal TEMP? - ambient Delta is 11-13C. New style fan filter even better, about 9-10C. Also reducing LTZ temperature would not help overall TC/ACAL performance as these effects are coming from different sources. Better and lower noise A1/A3 would be much more beneficial.
BR,
Illya
xDevs.com
On October 26, 2019 3:13:09 PM EDT, Philipp Cochems via volt-nuts volt-nuts@lists.febo.com wrote:
Hi,
one should always keep in mind that all the specifications are made to
achieve a certain yield in production. Assuming a Gaussian distribution
it
is not very unlikely, that we all have 3458As which perform much better
than specified. In that respect you cannot rely on vishay marketing.
Reducing the LTZ1000 temperature is also not that easy. The 3458A might
be
a metrology grade DMM, but it is also a system DMM. Even on my bench
the
TEMP? sometimes goes up to 47°C (at 26°C room) and it might be even
higher
in the summer. And it is not in a rack and only a HP 3488A is sitting
on
top which produces almost no heat. Running the 3458A in a rack full of
equipment will not allow to reduce the LTZ1000 that much.
I think the gap to the Fluke 8588A is too wide to catch up with only a
little bit improved specs. I think the main reason for the black
edition is
to ensure that many test rigs can work for the next years.
Best regards
Philipp
Am Sa., 26. Okt. 2019 um 20:48 Uhr schrieb Frank Stellmach <
frank.stellmach@freenet.de>:
@ Poul-Henning:
It's improbable that we'll see improved specs.
The presentation of the Black Edition already happened end of May,
and
the validation obviously was already running.
The description @ meilhaus already claims that no change in spec.
will
happen, and the linked Black Edition specs are really identical.
@ David and Bob:
RoHS is a European requirement only, and may not (yet) apply to other
markets.
KS obviously has stocked a bunch of lead containing units (they
called
it EU3458A) in their own premises before the exemption expired, and
as
this stock would be depleted later, they needed to design a lead
free
unit - initially for the EU market only.
I suppose they will sell the white version outside the EU for several
years, until their old component stock is depleted also.
I wonder what will happen to the EU34420A, which has the same
problem,
but lead free soldering might pose greater problems, due to e.m.f. @
nV
level.
@ Tom:
I can't tell anything about Golden Units, as I only own a regular,
white
HP3458A from 2001, which still performs better than specification..
+1.8ppm drift only, over 9 years for DCV, and Ohm is also much more
stable than specification, as far as I can judge.
Ok, the trick is, that it does not run 24/365, and my LTZ reference
board is set to ~ 65°C.
Concerning the internal A/D converter clock, there are several papers
about AC metrology, especially on AC power measurements, where the
PTB
improved the accuracy greatly by optically linking the internal clock
to
the rear of the 3458A, to mitigate the 10 or 100ns (?) clock phase
jitter. See also Rado Lapuh, Sampling With 3458A, page 149.
A practical schematic is not published, so it's very appreciated that
you provide one, but a description, how to apply this clock would be
necessary.
Frank
volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.
BR,
Illya Tsemenko
Apropos: I'm very interested to hear what REV? returns on the black edition.
--