Dave,
at first, all instruments need a stable temperature environment,
otherwise ppm measurements are meaningless,
I recommend setting up the lab in the basement, if available.
The 3458A is stable after about 2h, or 4h at most.
Check its inernal temperature, and use ACAL DCV frequently.
As the 3458A and the 4808 agree to <1ppm on the 10V range, there's good
reason that the basic DCV calibration of the 3458A is still fine. Both
these instruments do not drift that much if they are old vintage, and if
they are not powered on 24/7/365.
If the 3458A was unpowered most of the time, its reference very
probabaly did not drift at all.
If your 732A agrees also on 10V within a few ppm, you would have another
fix point.
Maybe you can send in the 732A for comparison to another volt-nut, or
build a transportable 7,15xxxV reference with an LTZ1000, as described
in the eevblog thread.
Because the 3458A is an AUTOCAL instrument, the 1V, 100V readings are
precise to about the same level, as the 10V range.
The 3458A makes better ratio transfers than the 720A, butter latter is
good for linearity check, anyhow.
1kV depends on the 100:1 divider of your special 3458A instrument, it
can be precise as a few ppm only, but unprecise as 12ppm, as described
in the specification. For 1kV measurements, all instruments need at
least 1min stabilization before making measurements, due to power
induced temperature drift.
To use the 720A on 1kV, may not give better accuracy, either, again due
to the high power drift.
A 752A would be required, or another ACAL instrument, like the 5440A, or
the 5720A.
The references inside the 7081A may drift the most, so I would trust
them less.
For Ohm, you may want to follow TiNs proposals.
Frank
I believe that 3458A will be more stable if it is never turned off.
The references are cooked in. There was a problem when production was moved
offshore, the references were stored cold after they had been cooked into
spec. When these units were shipped they did not meat spec. There was a
memo stating that if left on for 6 months the references should return to
spec.
We stored all our 3458As powered up even if we did not anticipate using
them for months.
On Sun, Jul 9, 2017 at 11:24 PM, Frank Stellmach <frank.stellmach@freenet.de
wrote:
So I have:
So please how best to proceed from here?
Thanks Dave
Dave,
at first, all instruments need a stable temperature environment, otherwise
ppm measurements are meaningless,
I recommend setting up the lab in the basement, if available.
The 3458A is stable after about 2h, or 4h at most.
Check its inernal temperature, and use ACAL DCV frequently.
As the 3458A and the 4808 agree to <1ppm on the 10V range, there's good
reason that the basic DCV calibration of the 3458A is still fine. Both
these instruments do not drift that much if they are old vintage, and if
they are not powered on 24/7/365.
If the 3458A was unpowered most of the time, its reference very probabaly
did not drift at all.
If your 732A agrees also on 10V within a few ppm, you would have another
fix point.
Maybe you can send in the 732A for comparison to another volt-nut, or
build a transportable 7,15xxxV reference with an LTZ1000, as described in
the eevblog thread.
Because the 3458A is an AUTOCAL instrument, the 1V, 100V readings are
precise to about the same level, as the 10V range.
The 3458A makes better ratio transfers than the 720A, butter latter is
good for linearity check, anyhow.
1kV depends on the 100:1 divider of your special 3458A instrument, it can
be precise as a few ppm only, but unprecise as 12ppm, as described in the
specification. For 1kV measurements, all instruments need at least 1min
stabilization before making measurements, due to power induced temperature
drift.
To use the 720A on 1kV, may not give better accuracy, either, again due to
the high power drift.
A 752A would be required, or another ACAL instrument, like the 5440A, or
the 5720A.
The references inside the 7081A may drift the most, so I would trust them
less.
For Ohm, you may want to follow TiNs proposals.
Frank
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/volt-nuts
and follow the instructions there.
--
John Phillips
Frank and others,
1kV depends on the 100:1 divider of your special 3458A instrument, it can be precise as a few ppm only, but unprecise as 12ppm,
as described in the specification.
The 3458A readings for all ranges except 1000V are now within the spec of the Datron 4808 (between 2.8ppm and 5.6ppm depending on range)
On the 1000V range for voltages up to and including 700V the 3458A reading are consistently low low but within the Datron spec of +/-5.6ppm
For voltages of 800V and above the 3458A reading starts off just within or close to spec of the Datron of 5.6ppm, but then falls away as the divider warms up (I think).
Here are the settling levels:
800V is 799.9946V 7.3 to 7.4ppm low
900V is 899.9901V 9.9ppm low
1000V is 999.9886V 11.2ppm low
Whether it is the Datron or the 3458A I'm not yet sure. That's why I'm doing the 720A alignment so I can cross check.
Dave
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Frank Stellmach
Sent: 10 July 2017 07:25
To: volt-nuts@febo.com
Subject: [volt-nuts] Guidance requested
So I have:
So please how best to proceed from here?
Thanks Dave
Dave,
at first, all instruments need a stable temperature environment, otherwise ppm measurements are meaningless, I recommend setting up the lab in the basement, if available.
The 3458A is stable after about 2h, or 4h at most.
Check its inernal temperature, and use ACAL DCV frequently.
As the 3458A and the 4808 agree to <1ppm on the 10V range, there's good reason that the basic DCV calibration of the 3458A is still fine. Both these instruments do not drift that much if they are old vintage, and if they are not powered on 24/7/365.
If the 3458A was unpowered most of the time, its reference very probabaly did not drift at all.
If your 732A agrees also on 10V within a few ppm, you would have another fix point.
Maybe you can send in the 732A for comparison to another volt-nut, or build a transportable 7,15xxxV reference with an LTZ1000, as described in the eevblog thread.
Because the 3458A is an AUTOCAL instrument, the 1V, 100V readings are precise to about the same level, as the 10V range.
The 3458A makes better ratio transfers than the 720A, butter latter is good for linearity check, anyhow.
1kV depends on the 100:1 divider of your special 3458A instrument, it can be precise as a few ppm only, but unprecise as 12ppm, as described in the specification. For 1kV measurements, all instruments need at least 1min stabilization before making measurements, due to power induced temperature drift.
To use the 720A on 1kV, may not give better accuracy, either, again due to the high power drift.
A 752A would be required, or another ACAL instrument, like the 5440A, or the 5720A.
The references inside the 7081A may drift the most, so I would trust them less.
For Ohm, you may want to follow TiNs proposals.
Frank